Hello all, some reporting and some questions, especially to the laser specialists around here ! I have been playing around with the setup I described in post #74 above, improving alignment in order to increase the signal. It became quickly a (so far unsuccessful) chase for the origin of the strange lines I get at 537nm and 543nm. But first the findings related to the acquisition of the raman spectra themselves:
- there seems to be no way in my setup to remove the broad background intensity on which the raman peaks are superimposed (see spectra I attached in posts #74 and 78 above). I suspect it is due to fluorescence in the samples themselves, since the background is very sample dependent. This brought background "bump" and the raman peaks seem to scale in a similar way, i.e. the shape of the spectrum does not change very considerably with increasing laser intensity, apart from an overall scaling. The signal to noise is better for intenser laser illumination (as expected). Is this in line with your experiences as well ?
- If re-using the BW TECH hardware, the orange fiber is definitely better to connect the output to the spectrometer. Due to its larger diameter, the signal is definitely better.
- As chloderic mentioned a few posts back, the long pass filters only work well under normal incidence +/- a few degrees. This means that 532nm straylight, that is scattered by off-axis parts of the setup, can still pass through the longpass filter (due to its angle with respect to the optical axis) and possibly make its way to the spectrometer.
- the use of 2 longpass filters in series does not seem to cause any specific problems. The combination blocks almost completely all 532nm light ! For my setup, using only one filter results in a saturated 532nm peak (but it is not causing any problems).
- it seems my resolution for raman lines is mainly driven by the laser monochromacity (or lack of it). I could not resolve some double lines, which are clearly resolved in some spectra I found on the internet acquired with home-made setups.
Now about the strange peaks at 537nm and 543nm: I tried almost everything I can think of to get rid of them, but without success:
- shining the laser directly into the setup, without fiber coupling: no difference (apart from some increase in laser intensity, no impact on the presence of the lines). Therefore, these peaks do not come from the fiber to couple the laser into the setup and I do not seem to have a problem with the fiber over-load mentioned by chloderic in post 76.
- blocking off-axis straylight makes not difference for these peaks. I have tried several simple (i.e. paper) internal baffles in the setup without success. I am more and more convinced the peaks come from on-axis light.
- putting the optical surfaces under a small angle, to avoid multiple reflections between parallel optical surfaces: no difference
- removing one or the other long-pass filter: no difference, except that the peaks become more intense (less blocked by the filters... note that the 543nm peak is beyond the blocking range of the filter !)
- beam dump : shielding the beam dump off from the optical path: no difference.
The only items I did not change are:
- the fiber to the spectrometer, but due to the low intensity of the light in that fiber, I have difficulties to believe that the light would trigger some raman peaks in the fiber itself, especially considering that for the double filter setup, the 532nm peak is almost completely absent).
- the lens on the sample side: I do not have any other suitable lens for that part, but I have difficulties to see what could cause such peaks in that lens (see below for more observations related to these peaks).
- the laser
Therefore my questions:
- any other ideas ?
- for the laser specialists here: could these lines be coming directly from the laser ? It is a simple 532nm dpss laser (no single mode) and I do not use any laser line filter. Can the lines be so far of the center 532nm line ? In addition, I see that these lines are extremely variable in intensity in time: one moment they are completely gone, a few moments later they saturate the spectrometer detector, without any change to the setup ! In the beginning I was puzzled about the fact that the 537nm and 543nm lines were brighter than the 532nm line, until I realised that the long-pass filter was not blocking the 543nm (and less efficient in blocking the 537nm light).
Thanks for any advice !