measurement and phone with ophir support helps me a little (unofficially :-D)
the only thing that is the same on all custom sensors with 6pin molex (95%)
1 - custom
2 - GND
3 - Vin (0.5 means that the sensor has clibrated 0.5V/1W - for 20W we need 10V + 2V => 12V, for 2W it is working with 6V)
4 -Analog Out
5 - custom
6 - custom
when i connect 12V power supply, the analog output works but the measurement confused me (~20mV random Steps)
0mW = -9mV, or -22mV
55mW = 0mV, or 18mV
103mW = 46mV or 68mV
2204mW = 1195mV or 1208mV or 1227mV or 1239mV
exact values between this steps are not possible. working with my selfmaded 10mm TEC LPM is more accurate :-D
the input laser are adjustet with my calibrated 20C-UAU USB-Head and StarCom. This is my reference head.
pin 1 has always ~5V output, when power is connectet, pin5 and 6 = 0V
Yeah doing some reasearch in our files and on line..
my initial thoughts are not correct.
S_l is correct.. the Datasheet shows the pinouts..
http://www.ophiropt.com/laser--measurement/sites/default/files/20C-SH_20C-UAS_20C-UAU.pdf
The .5V/W refers to the Analog output.. it
means that for every 0.5 volts measured by
the head sensor it will represent 1 Watt of
Laser Power.
I would put 12VDC on pin #3 and Ground on
Pin #2. Then between ground and pin #$ I
would put a good quality DMM to read the
Analog voltage with a 1 Watt beam on the
Sensor..
If all is as expected you will get a reading
of 0.500 Volts on the DMM.
Using any other Laser power on the Sensor
Head just multiply the reading by 2 to get
the mW of that measurement.
Remember the data sheet states the head
does not measure correctly under 10mW.
Or.. try decoding the RS232 Signals on Pin#6.
If that does not give accurate results then
perhaps the Amp board has been damaged
and is defective.
Jerry