Nice job on building your LPM.
I would make a few suggestions for more accurate readings....
1) the OPHIR Head should be completely out of the enclosure
to have all sides at ambient temperature equilibrium.
2) Pictures #8, #12,#16 and #17 show the Laser's beam profile
OFF the Sensor's active area center. You will get erroneous power
readings if not centered. The OPHIR heads are calibrated by OPHIR
at the exact center of the sensor.
3) In picture #9 you are WAY too close to the Sensor to get an
accurate reading. At that distance and Laser Power you are no
doubt adding radiated heat from the Laser LD and Laser head
to the actual Laser Power reading.
4) the cover was nicely designed but goes back to suggestion 1).
Jerry
Hi Jerry,
for my answers of the above:
1. I tried to expose as much of the head to ambient as possible. Even having to inset the side panel which is plastic to expose the back side to outside air. The rest are attached to the DIY aluminum chassis which also acts as an additional heatsink to ambient
2. There's no explanation there... other than it's hard to aim the laser using one hand while holding the cam on the other and looking through the camera screen.
3. There was no lens so that was the only way to put as much uncollimated light into the sensor as possible. The laser was from a cold start and had barely enough time to heat up when the pic was taken so there should be minimal effect from the heat. Either way, that was just for curiosity's sake rather than getting an exact power reading. I bet some of the uncollimated light is hitting the inside of the copper module rather than the absorber.
4. The cover is made from aluminum so it also acts as additional heatsinking to ambient which is suggested by Ophir (heatsinking for higher power handling).
But I do appreciate the suggestions and will take them into consideration when taking future measurements. :thanks: