Welcome to Laser Pointer Forums - discuss green laser pointers, blue laser pointers, and all types of lasers

LPF Donation via Stripe | LPF Donation - Other Methods

Links below open in new window

ArcticMyst Security by Avery

Review of the Ocean Optics USB2000 Spectrometer

Joined
Nov 1, 2006
Messages
549
Points
43
USB2000 SPECTROMETER

This is a long page with at least 25 images on it; dial-up users please allow for plenty of load time.


USB2000 Spectrometer, retail $3,205.00* (www.oceanoptics.com...)

Manufactured by Ocean Optics (http://oceanoptics.com)
Last updated 09-26-18


usb2000.jpg

what.gif


*** VERY, VERY IMPORTANT!!! ****
This product does not emit light of its own, so the standard review format will not be used and the product will be assigned a rating at once. A very high rating too, as during the time I've had it, it has thus far performed flawlessly.
rainbow.gif


Also, you must note that a lot of the features of the spectrometer and its software suite go unused, and therefore, unmentioned.

The USB2000 Spectrometer is a very small device (fits in the palm of your hand!) that can provide the spectral content of any UVC, UVB, UVA, visible, or NIR light source (ranging from ~175nm to ~874nm) , using a program called SpectraSuite to show the spectrum on your computer screen in graph format.

It is very important that you install the software on your computer BEFORE plugging the spectrometer in; even more so with the USB version!!!
The USB version plugs directly into any free USB port on your computer.

usb20002.gif

Here's a screen dump (yes it's really called that) of the SpectraSuite software window.

This program does a lot, but for this website sakes, I have just been taking screen dumps (yes, they're really called that), cropping them so only the spectral window is displayed, saving them in .GIF format (so that they're "web friendly"), and displaying them on this website; like this shining example:

alsproto.gif

This is the spectrum from an Arc LS Prototype flashlight.

To use the spectrometer, run SpectraSuite first. Once that is running, turn off the lights in the test area until you see a flat red line on the spectral window in SpectraSuite, position the light source to be measured near the port on the side of the spectrometer's body (where you see an indentation in the spectrometer's case), and move it closer to or farther away from this port until the highest point of the spectrum on your computer screen reaches the "4000" level or below. If the light source is too close, the tops of the spectrum will reach the top of the spectral window in SpectraSuite and flatten out - this is your cue to back up the light source so that the light entering the port on the side of the USB2000's body becomes dimmer.

At this point, I just hit the "Print Screen" key on my computer, open a graphics program like IrFanView and then "paste" (CTRL-V)) my screen capture into it (depending on the image viewer software you choose to use, you may ***OR*** may not have to first load any convienient image into it beforehand and then "paste" the screen dump over it). I then crop it so only the spectral window is displayed, save it in .GIF format (so that it's "web friendly"), and display it on this website.

I'm certain there are other methods of data capture, but for my purposes, this is more than sufficient. Like I mentioned earlier, the SpectraSuite program is *FAR* more versatile, but this method works just perfectly for me.
 





Joined
Nov 1, 2006
Messages
549
Points
43
TEST NOTES:
Test unit was loaned to me by TWO-CUBED.COM and was received on 03-06-06.

light.jpg

This picture shows you the relationship of visible light to UV and IR radiation.
Picture courtesy of a website fan and was used with permission.


UPDATE: 05-07-06
I have packed the USB2000 up so it can travel with me when I move to Sacramento around 05-22-06.
So I do not currently have it at my disposal for performing spectrographic analyses.


UPDATE: 07-06-06
The spectral chart is about as simple as it gets: wavelength starting at the left from ~175nm in the UVC portion of the spectrum to ~875nm in the near-infrared (NIR) region of the spectrum. The visible range is from ~380nm to ~780nm.
The higher the red line on the chart goes, the more radiation is being generated by the light source under test.

Let's take a look at one of these charts and see if I can tell you what's going on...

fl62006.gif

This is the spectrum from a typical Nichia NSPW500CS phosphor white LED, generally considered to be the brightest 5mm white LED as of early- to mid-2006.

As the chart shows, there is a strong peak in the blue at ~460nm, and a broad emission peaking at ~560nm in the yellow-green; overall, a cool white color is produced.



UPDATE: 02-01-07
The SpectraSuite software has been revised/updated; now the visible part of the spectrum can be colored, as shown in this example:

p1d.gif

This is the spectrum of a Fenix P1D CE flashlight that uses a high-powered Cree white LED.
I used a function of SpectraSuite to limit the spectral display to the visible wavelength range (385nm to 780nm).

UPDATE: 09-18-07
The person who loaned me the instrument now needs it back, so I no longer have it at my disposal.


UPDATE: 09-18-07
No, you aren't seeing things.
Yes, a same-day update.
I have been informed that this spectrometer or another similar model will be loaned to me sometime in the future; so spectroscopy will not be forever gone from this website.


UPDATE: 03-08-08
I now have permanent custody of the USB2000 spectrometer, with much thanks & gratitude to P.L. of TWO-CUBED!!!
rainbow.gif



UPDATE: 06-10-08
The optical cable appears to have failed, however I can still take spectra, and the optical cable from my PC2000-ISA spectrometer may very well fit - however since it is packed in an unknown box for a move to take place in a couple of weeks, I cannot verify this. But like I said, I can still take spectra; the only thing here is that I have to keep the instrument itself where I can directly irradiate it with the light source under test rather than keeping it tucked away and using the optical cable for this purpose.


UPDATE: 06-11-08
I found the PC200-ISA spectrometer's optical cable last night, and it does indeed fit and work with this spectrometer...so I'm back in business as they say.

UPDATE: 06-26-08
The baseline *CAN* be lowered through SpectraSuite; I had been doing this with Photoshop until now (primarily with lasers).
Here, let's let the following graph do the talking here:

rectyel.gif

Note that there is no visible emission outside this yellow LED's native emission band.

ss1.gif

See that red arrow?
That's the field that needs to be changed here - to "190" in this case.
The specific parameter that needs to be set is the "Y-Axis Range".


UPDATE: 07-24-08
I have disconnected this instrument from the computer and screwed the metal "bung" over the light input aperture to help protect it from foreign matter. The next time this instrument will get used is sometime after 07-28-08 when I'm at my new location. I could be without internet access until as late as 09-01-08!!! Though this is highly unlikely, the probability is not mathematically zero.


UPDATE: 05-28-09
I have discovered a setting on the SpectraSuite window that's proven rather useful over the last several days...see where it reads "Electric dark correction" in the screen dump below?
I ticked (left-clicked my mouse pointer in) that box, and now the baseline sinks even more deeply. Instead of having to set it close to 200, I can now set it to 10 - this gives me greater range to play with, especially when dealing with weaker light sources.
usb20009.gif


The only "fly in the ointment" as it were is that the maximum vertical height of 4,096 is decreased by just under 200.
Here, let me demonstrate this with another screen dump; this time of the top of a white LED's spectrum:

usb2008.gif

See how the top portion appears "flattened" yet the spectrum isn't quite at the top of the scale? This though I *CAN* definitely live with...after all, this adjustment benefits weaker light sources, not stronger ones.

usb20010.gif

The USB2000 in quiescent mode; this "spectrum" shows only detector noise, which is very easily cancelled out.
As I just said, this is extremely easy to compensate for; it actually looks a whole lot worse than it actually is.
 
Joined
Nov 1, 2006
Messages
549
Points
43
UPDATE: 03-06-10
I have fielded a number of questions about how to use this instrument over the last several months, so I think it's time to show you.

1: Fire up the spectrometer (of course) -- you do this by launching SpectraSuite.

2: You'll see a screen like the one below.

ss1.gif

See the little "x" characters at the upper right of the "Data Sources" and "Data Views" boxes?
Click on them; the following screen shows what you should now see.

ss2.gif

Where it reads "Graph (A)", look just below and to the right of it for a number of icons. Click on the fourth one from the left (the one that looks like a magnifying lens with the number "213" in blue lettering in it).

ss3.gif

Type "380" in the first box (X-Axis Range) labelled "Minimum", type "780" in the box to its right labelled "Maximum", and type "21" in the box to the lower left under "Y-Axis Range" Leave the lower right one (that reads "4095") alone. Click "Apply", then "Close".

ss4.gif

Place your cursor (mouse pointer) anywhere in the chart, and right-click on it. The box you see above should then pop up. Click on the "Visible Spectrum" tab, select "Show Visible Spectrum Layer", click the "Clip Visible Spectrum to Trendline" radio button, then click the "Send Layer to Back" button. Finally, click "OK".

Click on the "Electric Dark Correction" box near the upper right of the window; a green check mark should appear in it when you're successful.
It does not matter when in the setup process this box is checked; so long as you check it before you take your first spectrum.

ss5.gif

Finally, you may irradiate the spectrometer's sensor aperture with the LED under test. Move it slightly if the spectrum gets a flat top (indicative of excessive light) or does not go high enough (indicative of insufficient light). When the spectrum looks good to you, click on the button with the two thick black vertical lines in it.

What you do with this at this point is up to you; I simply press the key on my computer labelled "PrtScrn/Sys Req", open an image viewer, and press CTRL-V (paste). I then crop it so that only the spectrum itself is shown, and save it in .GIF format (this is a web-friendly format that does not cause distortions from .JPEG compression).

There appears to be no facility to "save settings" (well, not that I've yet been able to find anyway), so this setup procedure has to be performed every time you launch SpectraSuite.


UPDATE: 07-23-11
I performed spectroscopy of sunlight with the soft red plastic "bung" that fits on the sensor to protect it from things like dirt, mud, dust rabbits, monkey pee, horse puckey, etc. during storage or transport.

To wit:

usb20011.gif






UPDATE: 11-10-11
The SpectraSuite software by Ocean Optics has a couple of rather prominent web pages.

1: www.oceanoptics.com/Products/spectrasuite.asp

2: spectroscopytv.com/2008/07/17/the-basics-of-spectrasuite/





UPDATE: 11-16-11
The calibration report (data sheet) for this particular spectrometer is at this link if you're interested in viewing it. This is a .PDF file; you'll need to have Adobe Acrobat installed in order to view it.

I was furnished with it by N.D. from Ocean Optics on 11-15-11 after asking for it late on the evening of 11-14-11 by request of a website viewer who required spectroscopy of his Zilla "Reptile" Fluorescent Bulb. I really do appreciate the promptness of N.D. from Ocean Optics in getting this to me!!!
happy14.gif
Thank you ever so much!!!
thumbsup.gif




UPDATE: 02-06-16
The final release of SpectraSuite software is at http://oceanoptics.com... (approx. 80% down the web page). I've been using it for the last several months, and it has thus far performed flawlessly.

For the last few years, I've also been capturing spectral data files (in .TXT format that can be imported directly into Microsoft Excel or other spreadsheet software) that many of this website's visitors have requested.



UPDATE: 09-26-18
I have obtained the wavelength calibration data sheet; it's available right here.
It is equipped with Grating #2 (600 lines/mm), it has a 25µ slit, and its optical resolution (FWHM) is ~1.42nm.

I really do appreciate the promptness of N.D. from Ocean Optics in getting this data me (YES, A SECOND KUDOS TO NORM!)
happy14.gif
Thank you ever so much!!!
thumbsup.gif





fire.gif


MANUFACTURER: Ocean Optics
PRODUCT TYPE: Spectrometer
LAMP TYPE: N/A
No. OF LAMPS: N/A
BEAM TYPE: N/A
SWITCH TYPE: N/A
CASE MATERIAL: Aluminum
BEZEL: N/A
BATTERY: N/A
CURRENT CONSUMPTION: Unknown/unable to measure
WATER RESISTANT: No
SUBMERSIBLE: NO WAY HOZAY!!!
ACCESSORIES: Software CDs, connecting cable
SIZE: 90mm W x 63.5mm H x 35mm T (incl. feet)
WEIGHT: 220.20g (7.77 oz.)
COUNTRY OF MANUFACTURE: USA
WARRANTY: Unknown/not stated

PRODUCT RATING:

st5.gif
 

DTR

0
Joined
Jun 24, 2010
Messages
5,684
Points
113
Great thorough review as always.🍺

Started eyeballing some spectrometers recently so I can add some more details to my testing threads. Thanks for all the details. Lot of great info to absorb when I get a chance to go over it all in detail to get familiar some footing on them. Should motivate and give some direction. Thanks for posting.;)
 
Joined
Sep 20, 2013
Messages
17,426
Points
113
Too bad you had to send this unit back. I would suggest you use an Ocean Optics fiber optic cable to couple light into the spectrometer as this makes it much easier to get a clean source into it as opposed to using the slit alone.
 
Joined
Nov 1, 2006
Messages
549
Points
43
Too bad you had to send this unit back. I would suggest you use an Ocean Optics fiber optic cable to couple light into the spectrometer as this makes it much easier to get a clean source into it as opposed to using the slit alone.

If you read the 03-08-08 update, I have permanent custody of the instrument! :)
Here, let's snip that out of my review and post it here...BRB...

-------------------------------------------------------------------------------------------------------------------------------------------------------------
UPDATE: 03-08-08
I now have permanent custody of the USB2000 spectrometer, with much thanks & gratitude to P.L. of TWO-CUBED!!!
rainbow.gif

-------------------------------------------------------------------------------------------------------------------------------------------------------------

I agree that using the fiber optic cable as a coupler would be best, but it became broken in 2009 so I have to make do with irradiating the slit either directly (with weaker sources) or off-axis (with stronger sources).
 
Joined
Sep 20, 2008
Messages
17,622
Points
113
You never let us down with your in depth reviews...
Another excellent one. (y)

Jerry
 

kecked

0
Joined
Jun 18, 2012
Messages
950
Points
63
I have a similar newer unit also called usb2000 it’s does 200nm-1150nm. Maybe you can get it upgraded. Resolution is 1.2nm solid unit I love it. I used it our to 1150nm. And out to 266nm. Can’t say more on edges asi have no sources. It does see the 196nm mercury line or so I forget exact line. Love it love it love it.
 
Joined
Nov 21, 2009
Messages
318
Points
28
Congratulations on the epic new instrument. I cannot think of anyone more deserving!
I too, remember the time before blue LED's existed. And visiting the LED museum on dial up.
Now I have a cable modem and 2 flashlights over 5000 lumens, A BLF Q8 and a Sofirn Q8.
Best wishes from a long time reader and fan!
 

kecked

0
Joined
Jun 18, 2012
Messages
950
Points
63
On thing I need to figure is how to,tell the software to ignore a range. For example looking at Raman. Ignore the excitation wavelength. Overlords the detector so tell it to ignore a pixel range.
 




Top