LBII
The set up with my stable 808nm laser. (The same wavelength used to calibrate the laserbee's IIRC)
The Ophir reading (Minus 7 mV for zeroing)
The laserbee II reading:
HLPM
I also received a used HLPM recently.
This test was done with a pen laser as I don't have a lab style 405nm laser.
Ophir reading: 45mV (minus 7 for zeroing) so 38mW.
The Laserbee II read 35mW.
The HLPM read 4.4mV, and according to the correction factor chart it should be multiplied by 23.3. 102.52mW
That's 70% off!
When reading in the red spectrum it's fine, but I think the wavelength chart may be a little off.
Could any other users with an HLPM, a 405nm and another power meter confirm the results?
The HLPM's wavelength chart is most likely off, meaning that readings not in the 600-700nm spectrum will be off, I would like someone else to confirm this as my unit could be defective.
I'll respond to this
HLMP II 70% off accusation here....
1) First you did not buy your HLPM II brand new from us.
2) We have had 7 iterations of our Optical LPM Module starting
with the 30mW then 60mW... 120mW... 200mW then the HLPM II
series of which we had 3 upgrades. Each of which had their own
Optical correction chart due to us finding better sensors. Only
the last 2 HLPM II iterations could test 405nm wavelengths.
We had sent out HLPM IIs and test sheets to a number of members
here to test their 405nm Lasers in the past and their data was used
to help build the Charts.
3) We have no idea when your HLPM was manufactured nor which
model it is. That could be found if we knew who purchased it and
when.
4) We have no idea if the Optical Sensor was damaged by the
previous owner. This can easily be done by applying to much
power to the sensor
5) is the chart that you are using the one supplied with that soecific
HLPM II ??
6) Have you tested your DMM as outlined in the User Instructions
with the supplied resistors to see if you require the Shunt ??
Without this information you could easily be testing an older HLPM II
that was not designed to read the 405nm wavelengths.
If I am correct then the laserbee is not a solid product since units such as mine, and danefex's have different curve adjustments which are not accounted for and they will end up being off by up to 15%.
As to your other claims about your (now ours) LaserBee II
inaccuracy claims......
With today... we will have now spent 2 days trying to find out
where your now Claimed possible 15% Low readings on your
LaserBee II may actually come from.
I will be opening our own Thread to post our tests... results
and conclusions on your accusations of our LaserBee products
inaccuracy...
Coming soon.....
The sensor coating could also not be spectrally flat... he's said it is, but he's lied about the attributes of LaserBee products before, so it might not be.
Trevor
I must have missed that... :thinking:
Care to offer up a link......
Jerry
You can contact us at any time on our Website: J.BAUER Electronics