Re: B&Wtech 473nm unit / Spectrometer Mods & Info
Transmission data in "counts" for BWTEK BRP473 probe assembly when looking at a Tungsten Source:
This is black box connected with two fibers when you buy a "Blue Laser With Spectrometer ) assembly on Ebay. This is transmission data off collection port, not the 473 nm Launch Port on the side,
Note it cuts on at about 490 nm and there are narrow but deep transmission dips centered at 545 nm and 563.5 nm with a first maxima at 527 nm and the peak maxima is centered on 579 nm. Data taken with an Ocean Optics USB 2000 During laser calibration of a B&W Tek at work... Please see attached screen shot thumbnail way down at the bottom . (credit to my employer for owning the USB 2000)
Steve
Capture_1
191.128 -153.17
191.52 -2.17
191.912 1.83
192.304 -1.17
192.695 -0.17
193.087 30.83
193.479 16.83
193.87 4.83
194.262 8.83
194.653 8.83
195.045 12.83
195.436 6.83
195.828 23.83
196.219 9.83
196.611 5.83
197.002 4.83
197.393 4.83
197.785 15.83
198.176 5.83
198.567 5.83
198.958 7.83
199.349 10.83
199.74 4.83
200.131 23.83
200.523 6.83
200.914 4.83
201.304 7.83
201.695 13.83
202.086 10.83
202.477 3.83
202.868 7.83
203.259 10.83
203.65 10.83
204.04 15.83
204.431 9.83
204.822 17.83
205.212 3.83
205.603 16.83
205.993 12.83
206.384 8.83
206.774 8.83
207.165 8.83
207.555 8.83
207.946 17.83
208.336 46.83
208.726 14.83
209.117 17.83
209.507 10.83
209.897 12.83
210.287 11.83
210.678 9.83
211.068 10.83
211.458 11.83
211.848 11.83
212.238 4.83
212.628 9.83
213.018 2.83
213.408 41.83
213.798 15.83
214.187 10.83
214.577 10.83
214.967 11.83
215.357 4.83
215.747 11.83
216.136 9.83
216.526 11.83
216.915 16.83
217.305 13.83
217.695 12.83
218.084 7.83
218.474 9.83
218.863 12.83
219.252 10.83
219.642 8.83
220.031 13.83
220.42 12.83
220.81 10.83
221.199 8.83
221.588 16.83
221.977 10.83
222.366 14.83
222.756 8.83
223.145 12.83
223.534 8.83
223.923 9.83
224.312 7.83
224.7 8.83
225.089 10.83
225.478 12.83
225.867 3.83
226.256 11.83
226.645 7.83
227.033 7.83
227.422 11.83
227.811 10.83
228.199 6.83
228.588 20.83
228.976 12.83
229.365 14.83
229.753 7.83
230.142 22.83
230.53 13.83
230.919 5.83
231.307 6.83
231.695 14.83
232.084 13.83
232.472 15.83
232.86 21.83
233.248 13.83
233.636 10.83
234.024 8.83
234.412 8.83
234.8 13.83
235.188 27.83
235.576 12.83
235.964 9.83
236.352 9.83
236.74 11.83
237.128 2.83
237.516 13.83
237.903 10.83
238.291 10.83
238.679 7.83
239.066 10.83
239.454 12.83
239.842 13.83
240.229 17.83
240.617 12.83
241.004 12.83
241.392 13.83
241.779 35.83
242.166 25.83
242.554 13.83
242.941 12.83
243.328 14.83
243.715 12.83
244.103 9.83
244.49 32.83
244.877 8.83
245.264 9.83
245.651 10.83
246.038 14.83
246.425 10.83
246.812 10.83
247.199 10.83
247.586 14.83
247.973 17.83
248.359 10.83
248.746 10.83
249.133 10.83
249.52 13.83
249.906 9.83
250.293 14.83
250.68 13.83
251.066 14.83
251.453 8.83
251.839 14.83
252.226 9.83
252.612 15.83
252.998 11.83
253.385 11.83
253.771 10.83
254.157 89.83
254.544 11.83
254.93 11.83
255.316 6.83
255.702 13.83
256.088 18.83
256.474 4.83
256.86 6.83
257.246 10.83
257.632 11.83
258.018 10.83
258.404 12.83
258.79 8.83
259.176 34.83
259.561 11.83
259.947 16.83
260.333 14.83
260.719 12.83
261.104 13.83
261.49 29.83
261.875 12.83
262.261 12.83
262.646 11.83
263.032 10.83
263.417 44.83
263.803 14.83
264.188 11.83
264.573 6.83
264.959 19.83
265.344 10.83
265.729 13.83
266.114 2.83
266.499 10.83
266.885 7.83
267.27 14.83
267.655 15.83
268.04 6.83
268.425 12.83
268.81 15.83
269.194 11.83
269.579 44.83
269.964 19.83
270.349 15.83
270.734 13.83
271.118 11.83
271.503 9.83
271.888 13.83
272.272 38.83
272.657 11.83
273.041 13.83
273.426 12.83
273.81 11.83
274.195 12.83
274.579 12.83
274.964 15.83
275.348 5.83
275.732 15.83
276.116 9.83
276.501 8.83
276.885 13.83
277.269 7.83
277.653 14.83
278.037 12.83
278.421 16.83
278.805 10.83
279.189 13.83
279.573 13.83
279.957 17.83
280.341 12.83
280.724 16.83
281.108 17.83
281.492 7.83
281.876 15.83
282.259 15.83
282.643 9.83
283.027 9.83
283.41 15.83
283.794 19.83
284.177 25.83
284.561 13.83
284.944 9.83
285.327 11.83
285.711 8.83
286.094 7.83
286.477 11.83
286.861 10.83
287.244 17.83
287.627 12.83
288.01 8.83
288.393 14.83
288.776 12.83
289.159 8.83
289.542 9.83
289.925 51.83
290.308 18.83
290.691 18.83
291.074 14.83
291.457 10.83
291.839 6.83
292.222 8.83
292.605 12.83
292.987 14.83
293.37 31.83
293.753 11.83
294.135 13.83
294.518 10.83
294.9 9.83
295.282 6.83
295.665 11.83
296.047 7.83
296.43 11.83
296.812 4.83
297.194 8.83
297.576 13.83
297.958 16.83
298.341 4.83
298.723 10.83
299.105 9.83
299.487 11.83
299.869 15.83
300.251 12.83
300.633 18.83
301.015 12.83
301.396 14.83
301.778 9.83
302.16 11.83
302.542 9.83
302.923 10.83
303.305 11.83
303.687 37.83
304.068 10.83
304.45 17.83
304.831 6.83
305.213 14.83
305.594 12.83
305.976 10.83
306.357 6.83
306.738 16.83
307.12 8.83
307.501 8.83
307.882 8.83
308.263 17.83
308.644 7.83
309.026 15.83
309.407 16.83
309.788 8.83
310.169 16.83
310.55 12.83
310.931 11.83
311.311 10.83
311.692 9.83
312.073 17.83
312.454 18.83
312.835 15.83
313.215 16.83
313.596 19.83
313.977 10.83
314.357 20.83
314.738 9.83
315.118 15.83
315.499 19.83
315.879 21.83
316.26 5.83
316.64 11.83
317.02 8.83
317.401 8.83
317.781 16.83
318.161 26.83
318.541 8.83
318.921 12.83
319.301 9.83
319.682 19.83
320.062 3.83
320.442 10.83
320.822 10.83
321.201 15.83
321.581 16.83
321.961 7.83
322.341 12.83
322.721 13.83
323.1 16.83
323.48 11.83
323.86 13.83
324.239 10.83
324.619 11.83
324.999 11.83
325.378 13.83
325.758 14.83
326.137 6.83
326.516 16.83
326.896 11.83
327.275 12.83
327.654 12.83
328.034 11.83
328.413 11.83
328.792 10.83
329.171 13.83
329.55 12.83
329.929 9.83
330.308 13.83
330.687 20.83
331.066 21.83
331.445 16.83
331.824 10.83
332.203 14.83
332.582 13.83
332.96 9.83
333.339 68.83
333.718 12.83
334.096 9.83
334.475 4.83
334.854 10.83
335.232 9.83
335.611 12.83
335.989 4.83
336.367 15.83
336.746 11.83
337.124 10.83
337.502 13.83
337.881 6.83
338.259 13.83
338.637 32.83
339.015 13.83
339.393 12.83
339.771 15.83
340.149 11.83
340.527 15.83
340.905 10.83
341.283 14.83
341.661 11.83
342.039 12.83
342.417 53.83
342.795 11.83
343.172 10.83
343.55 12.83
343.928 15.83
344.305 11.83
344.683 14.83
345.06 6.83
345.438 9.83
345.815 19.83
346.193 18.83
346.57 8.83
346.948 12.83
347.325 14.83
347.702 8.83
348.079 11.83
348.457 45.83
348.834 10.83
349.211 10.83
349.588 13.83
349.965 10.83
350.342 15.83
350.719 10.83
351.096 11.83
351.473 14.83
351.85 21.83
352.226 39.83
352.603 16.83
352.98 16.83
353.357 14.83
353.733 15.83
354.11 10.83
354.487 5.83
354.863 14.83
355.24 11.83
355.616 13.83
355.992 11.83
356.369 10.83
356.745 11.83
357.122 19.83
357.498 10.83
357.874 13.83
358.25 7.83
358.626 6.83
359.003 10.83
359.379 8.83
359.755 10.83
360.131 32.83
360.507 12.83
360.883 9.83
361.258 13.83
361.634 12.83
362.01 12.83
362.386 15.83
362.762 14.83
363.137 12.83
363.513 6.83
363.889 6.83
364.264 9.83
364.64 14.83
365.015 10.83
365.391 10.83
365.766 11.83
366.142 8.83
366.517 15.83
366.892 11.83
367.267 11.83
367.643 16.83
368.018 18.83
368.393 15.83
368.768 39.83
369.143 17.83
369.518 18.83
369.893 18.83
370.268 12.83
370.643 12.83
371.018 19.83
371.393 14.83
371.768 18.83
372.142 23.83
372.517 19.83
372.892 31.83
373.266 15.83
373.641 16.83
374.016 18.83
374.39 13.83
374.765 18.83
375.139 24.83
375.514 26.83
375.888 11.83
376.262 25.83
376.637 21.83
377.011 22.83
377.385 21.83
377.759 20.83
378.133 23.83
378.507 32.83
378.882 33.83
379.256 25.83
379.63 23.83
380.003 26.83
380.377 23.83
380.751 29.83
381.125 27.83
381.499 25.83
381.873 28.83
382.246 27.83
382.62 24.83
382.994 27.83
383.367 25.83
383.741 31.83
384.114 35.83
384.488 33.83
384.861 31.83
385.235 22.83
385.608 31.83
385.981 32.83
386.355 18.83
386.728 25.83
387.101 33.83
387.474 28.83
387.847 27.83
388.22 36.83
388.593 33.83
388.966 36.83
389.339 34.83
389.712 31.83
390.085 33.83
390.458 30.83
390.831 28.83
391.204 34.83
391.576 29.83
391.949 28.83
392.322 29.83
392.694 29.83
393.067 31.83
393.439 33.83
393.812 31.83
394.184 39.83
394.557 47.83
394.929 39.83
395.301 220.83
395.674 31.83
396.046 34.83
396.418 33.83
396.79 36.83
397.162 38.83
397.534 32.83
397.906 41.83
398.278 38.83
398.65 42.83
399.022 31.83
399.394 40.83
399.766 41.83
400.138 35.83
400.51 40.83
400.881 34.83
401.253 35.83
401.625 26.83
401.996 25.83
402.368 28.83
402.739 25.83
403.111 35.83
403.482 21.83
403.854 25.83
404.225 25.83
404.597 25.83
404.968 46.83
405.339 36.83
405.71 41.83
406.081 47.83
406.453 47.83
406.824 47.83
407.195 48.83
407.566 41.83
407.937 43.83
408.308 45.83
408.679 52.83
409.049 44.83
409.42 52.83
409.791 57.83
410.162 46.83
410.532 52.83
410.903 51.83
411.274 52.83
411.644 55.83
412.015 51.83
412.385 53.83
412.756 50.83
413.126 41.83
413.496 41.83
413.867 47.83
414.237 36.83
414.607 37.83
414.977 34.83
415.348 32.83
415.718 54.83
416.088 31.83
416.458 32.83
416.828 38.83
417.198 43.83
417.568 47.83
417.938 49.83
418.308 54.83
418.677 60.83
419.047 70.83
419.417 73.83
419.786 74.83
420.156 74.83
420.526 71.83
420.895 72.83
421.265 66.83
421.634 72.83
422.004 73.83
422.373 77.83
422.742 81.83
423.112 90.83
423.481 109.83
423.85 96.83
424.219 85.83
424.589 80.83
424.958 80.83
425.327 74.83
425.696 72.83
426.065 82.83
426.434 68.83
426.803 60.83
427.172 61.83
427.54 51.83
427.909 48.83
428.278 59.83
428.647 73.83
429.015 67.83
429.384 81.83
429.753 70.83
430.121 100.83
430.49 80.83
430.858 96.83
431.226 93.83
431.595 97.83
431.963 101.83
432.331 98.83
432.7 93.83
433.068 94.83
433.436 88.83
433.804 85.83
434.172 97.83
434.54 97.83
434.908 76.83
435.276 87.83
435.644 90.83
436.012 78.83
436.38 82.83
436.748 89.83
437.116 94.83
437.483 127.83
437.851 123.83
438.219 106.83
438.586 107.83
438.954 105.83
439.321 101.83
439.689 104.83
440.056 103.83
440.424 109.83
440.791 104.83
441.158 101.83
441.526 95.83
441.893 94.83
442.26 91.83
442.627 74.83
442.994 90.83
443.361 96.83
443.728 113.83
444.095 132.83
444.462 127.83
444.829 153.83
445.196 138.83
445.563 139.83
445.93 122.83
446.296 108.83
446.663 96.83
447.03 84.83
447.396 82.83
447.763 70.83
448.129 64.83
448.496 63.83
448.862 72.83
449.229 73.83
449.595 88.83
449.961 98.83
450.328 97.83
450.694 110.83
451.06 101.83
451.426 79.83
451.792 81.83
452.159 61.83
452.525 57.83
452.891 45.83
453.256 34.83
453.622 31.83
453.988 39.83
454.354 28.83
454.72 26.83
455.086 28.83
455.451 21.83
455.817 29.83
456.183 32.83
456.548 31.83
456.914 31.83
457.279 33.83
457.645 31.83
458.01 34.83
458.375 29.83
458.741 26.83
459.106 31.83
459.471 20.83
459.836 29.83
460.202 22.83
460.567 27.83
460.932 20.83
461.297 18.83
461.662 32.83
462.027 23.83
462.392 15.83
462.757 16.83
463.121 20.83
463.486 17.83
463.851 20.83
464.216 20.83
464.58 16.83
464.945 22.83
465.309 18.83
465.674 21.83
466.038 17.83
466.403 19.83
466.767 18.83
467.132 31.83
467.496 18.83
467.86 16.83
468.224 20.83
468.589 23.83
468.953 17.83
469.317 18.83
469.681 19.83
470.045 28.83
470.409 19.83
470.773 23.83
471.137 22.83
471.501 26.83
471.864 55.83
472.228 51.83
472.592 10.83
472.956 22.83
473.319 37.83
473.683 23.83
474.046 22.83
474.41 30.83
474.773 22.83
475.137 23.83
475.5 28.83
475.864 22.83
476.227 27.83
476.59 21.83
476.953 25.83
477.317 25.83
477.68 33.83
478.043 47.83
478.406 25.83
478.769 25.83
479.132 25.83
479.495 26.83
479.858 31.83
480.221 24.83
480.583 29.83
480.946 24.83
481.309 29.83
481.671 25.83
482.034 30.83
482.397 27.83
482.759 31.83
483.122 26.83
483.484 26.83
483.847 28.83
484.209 34.83
484.571 36.83
484.934 35.83
485.296 42.83
485.658 41.83
486.02 64.83
486.382 76.83
486.744 97.83
487.106 126.83
487.468 164.83
487.83 213.83
488.192 263.83
488.554 319.83
488.916 402.83
489.278 459.83
489.639 484.83
490.001 517.83
490.363 561.83
490.724 587.83
491.086 597.83
491.447 635.83
491.809 676.83
492.17 702.83
492.532 765.83
492.893 792.83
493.254 888.83
493.616 986.83
493.977 1111.83
494.338 1213.83
494.699 1335.83
495.06 1426.83
495.421 1546.83
495.782 1621.83
496.143 1639.83
496.504 1722.83
496.865 1760.83
497.226 1796.83
497.587 1841.83
497.948 1889.83
498.308 1877.83
498.669 1905.83
499.03 1963.83
499.39 1945.83
499.751 1935.83
500.111 1953.83
500.472 2013.83
500.832 2007.83
501.192 1998.83
501.553 2029.83
501.913 2088.83
502.273 2055.83
502.633 2078.83
502.993 2088.83
503.354 2104.83
503.714 2133.83
504.074 2160.83
504.434 2147.83
504.794 2160.83
505.153 2190.83
505.513 2201.83
505.873 2199.83
506.233 2232.83
506.593 2217.83
506.952 2243.83
507.312 2261.83
507.671 2295.83
508.031 2298.83
508.39 2310.83
508.75 2330.83
509.109 2334.83
509.469 2362.83
509.828 2341.83
510.187 2385.83
510.547 2386.83
510.906 2385.83
511.265 2445.83
511.624 2404.83
511.983 2441.83
512.342 2479.83
512.701 2499.83
513.06 2475.83
513.419 2506.83
513.778 2514.83
514.136 2544.83
514.495 2544.83
514.854 2529.83
515.213 2583.83
515.571 2559.83
515.93 2597.83
516.288 2639.83
516.647 2618.83
517.005 2696.83
517.364 2665.83
517.722 2708.83
518.08 2732.83
518.439 2744.83
518.797 2771.83
519.155 2760.83
519.513 2816.83
519.871 2807.83
520.229 2847.83
520.587 2821.83
520.945 2840.83
521.303 2909.83
521.661 2938.83
522.019 2913.83
522.377 2911.83
522.735 2940.83
523.092 2956.83
523.45 2952.83
523.808 2984.83
524.165 2993.83
524.523 3012.83
524.88 3009.83
525.238 3016.83
525.595 3036.83
525.952 3031.83
526.31 3030.83
526.667 3058.83
527.024 3084.83
527.381 3077.83
527.738 3106.83
528.096 3119.83
528.453 3085.83
528.81 3111.83
529.167 3145.83
529.524 3104.83
529.88 3128.83
530.237 3058.83
530.594 2987.83
530.951 2954.83
531.307 2846.83
531.664 2756.83
532.021 2615.83
532.377 2459.83
532.734 2314.83
533.09 2149.83
533.447 2005.83
533.803 1868.83
534.159 1681.83
534.516 1587.83
534.872 1400.83
535.228 1285.83
535.584 1180.83
535.94 1048.83
536.297 956.83
536.653 892.83
537.009 816.83
537.365 735.83
537.72 696.83
538.076 652.83
538.432 593.83
538.788 564.83
539.144 515.83
539.499 484.83
539.855 453.83
540.21 432.83
540.566 411.83
540.922 400.83
541.277 388.83
541.632 381.83
541.988 364.83
542.343 351.83
542.698 344.83
543.054 333.83
543.409 340.83
543.764 340.83
544.119 358.83
544.474 338.83
544.829 337.83
545.184 342.83
545.539 337.83
545.894 359.83
546.249 377.83
546.604 382.83
546.958 399.83
547.313 423.83
547.668 449.83
548.022 492.83
548.377 498.83
548.731 547.83
549.086 599.83
549.44 665.83
549.795 713.83
550.149 777.83
550.503 860.83
550.858 946.83
551.212 1062.83
551.566 1170.83
551.92 1276.83
552.274 1407.83
552.628 1498.83
552.982 1628.83
553.336 1725.83
553.69 1830.83
554.044 1864.83
554.398 1930.83
554.751 1948.83
555.105 1915.83
555.459 1819.83
555.812 1744.83
556.166 1720.83
556.519 1538.83
556.873 1441.83
557.226 1369.83
557.58 1232.83
557.933 1147.83
558.286 1049.83
558.64 979.83
558.993 905.83
559.346 859.83
559.699 801.83
560.052 772.83
560.405 800.83
560.758 732.83
561.111 725.83
561.464 700.83
561.817 696.83
562.17 692.83
562.523 701.83
562.875 707.83
563.228 701.83
563.581 733.83
563.933 726.83
564.286 775.83
564.638 828.83
564.991 876.83
565.343 882.83
565.696 904.83
566.048 971.83
566.4 1010.83
566.752 1097.83
567.105 1160.83
567.457 1226.83
567.809 1303.83
568.161 1404.83
568.513 1500.83
568.865 1627.83
569.217 1738.83
569.569 1872.83
569.92 1983.83
570.272 2149.83
570.624 2288.83
570.976 2413.83
571.327 2550.83
571.679 2715.83
572.03 2847.83
572.382 3017.83
572.733 3121.83
573.085 3226.83
573.436 3281.83
573.787 3379.83
574.139 3424.83
574.49 3425.83
574.841 3426.83
575.192 3462.83
575.543 3476.83
575.894 3479.83
576.245 3443.83
576.596 3439.83
576.947 3400.83
577.298 3409.83
577.649 3395.83
578 3391.83
578.35 3389.83
578.701 3348.83
579.052 3401.83
579.402 3387.83
579.753 3394.83
580.103 3393.83
580.454 3425.83
580.804 3432.83
581.155 3438.83
581.505 3487.83
581.855 3488.83
582.205 3531.83
582.556 3492.83
582.906 3495.83
583.256 3495.83
583.606 3488.83
583.956 3476.83
584.306 3493.83
584.656 3482.83
585.005 3413.83
585.355 3403.83
585.705 3381.83
586.055 3311.83
586.404 3303.83
586.754 3314.83
587.104 3300.83
587.453 3214.83
587.803 3226.83
588.152 3189.83
588.502 3177.83
588.851 3107.83
589.2 3123.83
589.549 3113.83
589.899 3114.83
590.248 3099.83
590.597 3086.83
590.946 3115.83
591.295 3089.83
591.644 3166.83
591.993 3097.83
592.342 3112.83
592.691 3148.83
593.039 3147.83
593.388 3173.83
593.737 3183.83
594.086 3202.83
594.434 3216.83
594.783 3215.83
595.131 3238.83
595.48 3287.83
595.828 3271.83
596.177 3286.83
596.525 3311.83
596.873 3336.83
597.221 3338.83
597.57 3364.83
597.918 3397.83
598.266 3375.83
598.614 3380.83
598.962 3397.83
599.31 3384.83
599.658 3381.83
600.006 3416.83
600.354 3392.83
600.701 3407.83
601.049 3397.83
601.397 3361.83
601.744 3354.83
602.092 3328.83
602.439 3335.83
602.787 3322.83
603.134 3296.83
603.482 3262.83
603.829 3224.83
604.177 3241.83
604.524 3149.83
604.871 3134.83
605.218 3063.83
605.565 3095.83
605.912 3058.83
606.259 3003.83
606.606 2967.83
606.953 2983.83
607.3 2915.83
607.647 2876.83
607.994 2874.83
608.341 2828.83
608.687 2794.83
609.034 2785.83
609.381 2751.83
609.727 2706.83
610.074 2716.83
610.42 2657.83
610.767 2664.83
611.113 2655.83
611.459 2610.83
611.806 2623.83
612.152 2591.83
612.498 2637.83
612.844 2584.83
613.19 2577.83
613.536 2589.83
613.882 2544.83
614.228 2582.83
614.574 2539.83
614.92 2573.83
615.266 2547.83
615.612 2568.83
615.957 2556.83
616.303 2589.83
616.649 2572.83
616.994 2566.83
617.34 2571.83
617.685 2589.83
618.031 2556.83
618.376 2557.83
618.721 2564.83
619.067 2531.83
619.412 2554.83
619.757 2542.83
620.102 2517.83
620.448 2483.83
620.793 2494.83
621.138 2477.83
621.483 2434.83
621.827 2445.83
622.172 2437.83
622.517 2421.83
622.862 2384.83
623.207 2377.83
623.551 2366.83
623.896 2373.83
624.241 2304.83
624.585 2314.83
624.93 2299.83
625.274 2294.83
625.619 2252.83
625.963 2275.83
626.307 2217.83
626.651 2262.83
626.996 2241.83
627.34 2255.83
627.684 2257.83
628.028 2248.83
628.372 2322.83
628.716 2279.83
629.06 2333.83
629.404 2323.83
629.748 2373.83
630.091 2386.83
630.435 2429.83
630.779 2442.83
631.123 2481.83
631.466 2530.83
631.81 2562.83
632.153 2588.83
632.497 2635.83
632.84 2657.83
633.183 2719.83
633.527 2774.83
633.87 2804.83
634.213 2832.83
634.556 2874.83
634.9 2867.83
635.243 2907.83
635.586 2860.83
635.929 2873.83
636.272 2859.83
636.614 2860.83
636.957 2860.83
637.3 2808.83
637.643 2760.83
637.986 2711.83
638.328 2671.83
638.671 2647.83
639.013 2611.83
639.356 2557.83
639.698 2505.83
640.041 2468.83
640.383 2400.83
640.725 2338.83
641.068 2315.83
641.41 2235.83
641.752 2204.83
642.094 2131.83
642.436 2106.83
642.778 2070.83
643.12 2028.83
643.462 2013.83
643.804 2000.83
644.146 1943.83
644.488 1948.83
644.829 1908.83
645.171 1882.83
645.513 1915.83
645.854 1861.83
646.196 1878.83
646.537 1899.83
646.879 1893.83
647.22 1909.83
647.562 1934.83
647.903 1925.83
648.244 1994.83
648.585 2018.83
648.926 2066.83
649.268 2064.83
649.609 2124.83
649.95 2182.83
650.291 2249.83
650.632 2286.83
650.972 2326.83
651.313 2393.83
651.654 2438.83
651.995 2514.83
652.335 2536.83
652.676 2529.83
653.017 2588.83
653.357 2611.83
653.698 2610.83
654.038 2638.83
654.378 2554.83
654.719 2623.83
655.059 2524.83
655.399 2530.83
655.74 2480.83
656.08 2415.83
656.42 2406.83
656.76 2385.83
657.1 2336.83
657.44 2259.83
657.78 2199.83
658.12 2108.83
658.459 2111.83
658.799 2031.83
659.139 1962.83
659.479 1940.83
659.818 1900.83
660.158 1867.83
660.497 1796.83
660.837 1728.83
661.176 1694.83
661.516 1654.83
661.855 1619.83
662.194 1575.83
662.533 1570.83
662.873 1512.83
663.212 1509.83
663.551 1474.83
663.89 1439.83
664.229 1425.83
664.568 1424.83
664.907 1428.83
665.245 1412.83
665.584 1404.83
665.923 1407.83
666.262 1412.83
666.6 1455.83
666.939 1434.83
667.277 1466.83
667.616 1486.83
667.954 1527.83
668.293 1539.83
668.631 1564.83
668.969 1629.83
669.308 1711.83
669.646 1748.83
669.984 1803.83
670.322 1853.83
670.66 1944.83
670.998 1988.83
671.336 2042.83
671.674 2130.83
672.012 2177.83
672.35 2233.83
672.688 2267.83
673.025 2335.83
673.363 2335.83
673.701 2340.83
674.038 2362.83
674.376 2349.83
674.713 2323.83
675.051 2313.83
675.388 2269.83
675.725 2240.83
676.063 2232.83
676.4 2186.83
676.737 2138.83
677.074 2110.83
677.411 2071.83
677.748 2020.83
678.085 1972.83
678.422 1918.83
678.759 1918.83
679.096 1867.83
679.433 1822.83
679.769 1805.83
680.106 1752.83
680.443 1730.83
680.779 1742.83
681.116 1661.83
681.452 1647.83
681.789 1623.83
682.125 1611.83
682.461 1593.83
682.798 1589.83
683.134 1579.83
683.47 1580.83
683.806 1549.83
684.142 1586.83
684.478 1591.83
684.814 1624.83
685.15 1636.83
685.486 1638.83
685.822 1691.83
686.158 1702.83
686.494 1753.83
686.829 1819.83
687.165 1878.83
687.501 1899.83
687.836 1954.83
688.172 2031.83
688.507 2058.83
688.843 2099.83
689.178 2104.83
689.513 2167.83
689.848 2111.83
690.184 2129.83
690.519 2082.83
690.854 2087.83
691.189 2032.83
691.524 1973.83
691.859 1978.83
692.194 1920.83
692.529 1900.83
692.864 1862.83
693.198 1774.83
693.533 1732.83
693.868 1707.83
694.202 1679.83
694.537 1645.83
694.871 1613.83
695.206 1576.83
695.54 1536.83
695.875 1499.83
696.209 1437.83
696.543 1424.83
696.877 1373.83
697.212 1298.83
697.546 1294.83
697.88 1232.83
698.214 1199.83
698.548 1150.83
698.882 1106.83
699.215 1051.83
699.549 1022.83
699.883 979.83
700.217 937.83
700.55 900.83
700.884 871.83
701.218 832.83
701.551 801.83
701.885 783.83
702.218 744.83
702.551 724.83
702.885 703.83
703.218 677.83
703.551 656.83
703.884 644.83
704.217 634.83
704.551 618.83
704.884 600.83
705.217 594.83
705.549 593.83
705.882 564.83
706.215 573.83
706.548 558.83
706.881 557.83
707.213 564.83
707.546 539.83
707.878 564.83
708.211 553.83
708.543 564.83
708.876 563.83
709.208 566.83
709.541 568.83
709.873 576.83
710.205 578.83
710.537 598.83
710.869 583.83
711.201 597.83
711.533 604.83
711.865 678.83
712.197 648.83
712.529 684.83
712.861 683.83
713.193 715.83
713.524 743.83
713.856 751.83
714.188 785.83
714.519 797.83
714.851 848.83
715.182 874.83
715.514 893.83
715.845 926.83
716.176 958.83
716.507 1000.83
716.839 1045.83
717.17 1044.83
717.501 1104.83
717.832 1130.83
718.163 1161.83
718.494 1203.83
718.825 1226.83
719.156 1256.83
719.487 1314.83
719.817 1306.83
720.148 1332.83
720.479 1349.83
720.809 1369.83
721.14 1400.83
721.47 1415.83
721.801 1412.83
722.131 1436.83
722.461 1447.83
722.792 1433.83
723.122 1458.83
723.452 1466.83
723.782 1482.83
724.112 1466.83
724.442 1483.83
724.772 1473.83
725.102 1428.83
725.432 1444.83
725.762 1448.83
726.092 1437.83
726.422 1441.83
726.751 1390.83
727.081 1373.83
727.411 1388.83
727.74 1344.83
728.07 1345.83
728.399 1326.83
728.728 1294.83
729.058 1291.83
729.387 1257.83
729.716 1275.83
730.045 1210.83
730.375 1208.83
730.704 1187.83
731.033 1175.83
731.362 1143.83
731.691 1126.83
732.019 1089.83
732.348 1083.83
732.677 1071.83
733.006 1039.83
733.334 1013.83
733.663 1021.83
733.992 998.83
734.32 982.83
734.649 959.83
734.977 969.83
735.305 908.83
735.634 911.83
735.962 886.83
736.29 922.83
736.618 885.83
736.946 856.83
737.274 848.83
737.602 839.83
737.93 840.83
738.258 862.83
738.586 826.83
738.914 808.83
739.242 821.83
739.569 816.83
739.897 796.83
740.225 780.83
740.552 807.83
740.88 774.83
741.207 790.83
741.534 781.83
741.862 767.83
742.189 768.83
742.516 786.83
742.843 812.83
743.171 785.83
743.498 817.83
743.825 814.83
744.152 823.83
744.479 853.83
744.805 861.83
745.132 851.83
745.459 887.83
745.786 911.83
746.112 907.83
746.439 909.83
746.766 933.83
747.092 928.83
747.419 947.83
747.745 943.83
748.071 926.83
748.398 923.83
748.724 930.83
749.05 904.83
749.376 892.83
749.702 872.83
750.028 876.83
750.354 877.83
750.68 848.83
751.006 854.83
751.332 838.83
751.658 856.83
751.984 842.83
752.309 842.83
752.635 838.83
752.96 824.83
753.286 813.83
753.611 834.83
753.937 814.83
754.262 919.83
754.588 820.83
754.913 817.83
755.238 819.83
755.563 828.83
755.888 796.83
756.213 813.83
756.538 787.83
756.863 813.83
757.188 788.83
757.513 800.83
757.838 787.83
758.163 773.83
758.487 758.83
758.812 755.83
759.137 750.83
759.461 744.83
759.786 751.83
760.11 723.83
760.435 720.83
760.759 681.83
761.083 686.83
761.407 684.83
761.732 685.83
762.056 677.83
762.38 661.83
762.704 656.83
763.028 650.83
763.352 655.83
763.676 630.83
763.999 636.83
764.323 638.83
764.647 651.83
764.97 664.83
765.294 666.83
765.618 656.83
765.941 675.83
766.265 685.83
766.588 682.83
766.911 681.83
767.235 695.83
767.558 703.83
767.881 713.83
768.204 711.83
768.527 706.83
768.85 714.83
769.173 700.83
769.496 702.83
769.819 698.83
770.142 685.83
770.465 654.83
770.787 645.83
771.11 635.83
771.433 612.83
771.755 597.83
772.078 598.83
772.4 597.83
772.723 591.83
773.045 597.83
773.367 605.83
773.689 568.83
774.012 570.83
774.334 551.83
774.656 559.83
774.978 604.83
775.3 555.83
775.622 583.83
775.944 582.83
776.266 571.83
776.587 560.83
776.909 598.83
777.231 573.83
777.552 605.83
777.874 581.83
778.195 594.83
778.517 614.83
778.838 590.83
779.16 623.83
779.481 614.83
779.802 621.83
780.123 610.83
780.444 614.83
780.766 619.83
781.087 617.83
781.408 611.83
781.728 619.83
782.049 591.83
782.37 592.83
782.691 597.83
783.012 598.83
783.332 595.83
783.653 582.83
783.973 598.83
784.294 599.83
784.614 589.83
784.935 589.83
785.255 600.83
785.575 602.83
785.896 599.83
786.216 604.83
786.536 598.83
786.856 572.83
787.176 574.83
787.496 565.83
787.816 536.83
788.136 534.83
788.456 507.83
788.775 489.83
789.095 462.83
789.415 443.83
789.734 417.83
790.054 398.83
790.373 396.83
790.693 371.83
791.012 363.83
791.332 350.83
791.651 347.83
791.97 334.83
792.289 319.83
792.608 322.83
792.928 305.83
793.247 303.83
793.566 313.83
793.884 308.83
794.203 298.83
794.522 301.83
794.841 297.83
795.16 299.83
795.478 288.83
795.797 293.83
796.115 314.83
796.434 316.83
796.752 309.83
797.071 323.83
797.389 316.83
797.707 310.83
798.025 330.83
798.344 343.83
798.662 330.83
798.98 350.83
799.298 361.83
799.616 378.83
799.934 420.83
800.251 403.83
800.569 392.83
800.887 421.83
801.205 420.83
801.522 434.83
801.84 437.83
802.157 452.83
802.475 473.83
802.792 475.83
803.11 485.83
803.427 489.83
803.744 500.83
804.061 482.83
804.378 481.83
804.696 468.83
805.013 438.83
805.33 420.83
805.647 383.83
805.963 353.83
806.28 342.83
806.597 359.83
806.914 286.83
807.23 278.83
807.547 274.83
807.864 254.83
808.18 240.83
808.497 224.83
808.813 222.83
809.129 206.83
809.446 196.83
809.762 193.83
810.078 185.83
810.394 173.83
810.71 165.83
811.026 155.83
811.342 155.83
811.658 158.83
811.974 145.83
812.29 188.83
812.605 145.83
812.921 138.83
813.237 138.83
813.552 141.83
813.868 131.83
814.183 133.83
814.499 135.83
814.814 132.83
815.129 125.83
815.445 131.83
815.76 144.83
816.075 131.83
816.39 133.83
816.705 133.83
817.02 134.83
817.335 122.83
817.65 130.83
817.965 151.83
818.28 149.83
818.594 149.83
818.909 142.83
819.224 147.83
819.538 158.83
819.853 153.83
820.167 156.83
820.481 149.83
820.796 163.83
821.11 170.83
821.424 163.83
821.739 177.83
822.053 181.83
822.367 183.83
822.681 196.83
822.995 199.83
823.309 192.83
823.622 210.83
823.936 220.83
824.25 218.83
824.564 227.83
824.877 226.83
825.191 230.83
825.504 230.83
825.818 226.83
826.131 218.83
826.445 221.83
826.758 207.83
827.071 203.83
827.384 188.83
827.698 184.83
828.011 176.83
828.324 171.83
828.637 153.83
828.95 148.83
829.262 126.83
829.575 124.83
829.888 117.83
830.201 117.83
830.513 114.83
830.826 91.83
831.138 98.83
831.451 92.83
831.763 88.83
832.076 84.83
832.388 82.83
832.7 80.83
833.013 71.83
833.325 71.83
833.637 68.83
833.949 68.83
834.261 69.83
834.573 72.83
834.885 101.83
835.197 72.83
835.508 68.83
835.82 65.83
836.132 68.83
836.443 66.83
836.755 63.83
837.066 76.83
837.378 60.83
837.689 62.83
838.001 67.83
838.312 66.83
838.623 68.83
838.934 61.83
839.245 72.83
839.556 68.83
839.867 80.83
840.178 71.83
840.489 67.83
840.8 75.83
841.111 79.83
841.422 93.83
841.732 66.83
842.043 81.83
842.353 74.83
842.664 79.83
842.974 71.83
843.285 70.83
843.595 75.83
843.905 72.83
844.216 80.83
844.526 68.83
844.836 65.83
845.146 56.83
845.456 55.83
845.766 54.83
846.076 48.83
846.386 55.83
846.696 39.83
847.005 52.83
847.315 49.83
847.625 48.83
847.934 46.83
848.244 50.83
848.553 38.83
848.863 46.83
849.172 38.83
849.481 36.83
849.79 49.83
850.1 36.83
850.409 40.83
850.718 40.83
851.027 42.83
851.336 36.83
851.645 41.83
851.954 46.83
852.262 47.83
852.571 41.83
852.88 38.83
853.188 71.83
853.497 48.83
853.806 45.83
854.114 43.83
854.422 45.83
854.731 47.83
855.039 45.83
855.347 43.83
855.655 53.83
855.964 72.83
856.272 44.83
856.58 48.83
856.888 43.83
857.196 43.83
857.503 41.83
857.811 37.83
858.119 44.83
858.427 45.83
858.734 44.83
859.042 40.83
859.349 43.83
859.657 49.83
859.964 42.83
860.272 50.83
860.579 46.83
860.886 53.83
861.193 46.83
861.5 46.83
861.807 37.83
862.115 43.83
862.421 41.83
862.728 44.83
863.035 41.83
863.342 57.83
863.649 46.83
863.955 47.83
864.262 44.83
864.569 45.83
864.875 43.83
865.182 40.83
865.488 42.83
865.794 50.83
866.101 43.83
866.407 45.83
866.713 55.83
867.019 45.83
867.325 43.83
867.631 45.83
867.937 51.83
868.243 55.83
868.549 46.83
868.855 44.83
869.16 41.83
869.466 47.83
869.772 48.83
870.077 53.83
870.383 50.83
870.688 49.83
870.994 76.83
871.299 48.83
871.604 45.83
871.909 49.83
872.215 44.83
872.52 50.83
872.825 50.83
873.13 47.83
873.435 49.83
873.74 46.83
874.044 47.83
874.349 52.83
874.654 53.83
874.959 49.83
875.263 50.83
875.568 46.83
875.872 46.83
876.177 51.83
876.481 52.83
876.785 51.83
877.09 53.83
877.394 56.83
877.698 49.83
878.002 51.83
878.306 48.83
878.61 47.83
878.914 55.83
879.218 51.83
879.522 45.83
879.825 56.83
880.129 56.83
880.433 51.83
880.736 51.83
881.04 52.83
881.343 78.83
881.647 52.83
881.95 54.83
882.253 53.83
882.557 51.83
882.86 42.83
883.163 90.83
883.466 51.83
883.769 49.83
884.072 51.83
884.375 46.83
884.678 51.83
884.981 50.83
885.283 50.83
885.586 47.83
885.889 58.83
886.191 50.83
886.494 52.83
886.796 55.83
887.099 51.83
887.401 55.83
887.703 54.83
888.005 47.83
888.308 48.83
888.61 47.83
888.912 44.83
889.214 50.83
889.516 48.83
889.818 45.83
890.119 52.83
890.421 48.83
890.723 55.83
891.024 53.83
891.326 45.83
891.628 48.83
891.929 48.83
892.231 51.83
892.532 53.83
892.833 50.83
893.134 80.83
893.436 53.83
893.737 48.83
894.038 43.83
894.339 43.83
894.64 73.83
894.941 42.83
895.242 54.83
895.542 53.83
895.843 47.83
896.144 48.83
896.445 62.83
896.745 58.83
897.046 47.83
897.346 43.83
897.646 46.83
897.947 53.83
898.247 51.83
898.547 49.83
898.847 87.83
899.148 101.83
899.448 44.83
899.748 46.83
900.048 52.83
900.347 49.83
900.647 43.83
900.947 47.83
901.247 42.83
901.546 49.83
901.846 41.83
902.146 120.83
902.445 46.83
902.744 43.83
903.044 45.83
903.343 40.83
903.642 45.83
903.942 42.83
904.241 42.83
904.54 44.83
904.839 44.83
905.138 46.83
905.437 36.83
905.736 40.83
906.034 44.83
906.333 45.83
Transmission data in "counts" for BWTEK BRP473 probe assembly when looking at a Tungsten Source:
This is black box connected with two fibers when you buy a "Blue Laser With Spectrometer ) assembly on Ebay. This is transmission data off collection port, not the 473 nm Launch Port on the side,
Note it cuts on at about 490 nm and there are narrow but deep transmission dips centered at 545 nm and 563.5 nm with a first maxima at 527 nm and the peak maxima is centered on 579 nm. Data taken with an Ocean Optics USB 2000 During laser calibration of a B&W Tek at work... Please see attached screen shot thumbnail way down at the bottom . (credit to my employer for owning the USB 2000)
Steve
Capture_1
191.128 -153.17
191.52 -2.17
191.912 1.83
192.304 -1.17
192.695 -0.17
193.087 30.83
193.479 16.83
193.87 4.83
194.262 8.83
194.653 8.83
195.045 12.83
195.436 6.83
195.828 23.83
196.219 9.83
196.611 5.83
197.002 4.83
197.393 4.83
197.785 15.83
198.176 5.83
198.567 5.83
198.958 7.83
199.349 10.83
199.74 4.83
200.131 23.83
200.523 6.83
200.914 4.83
201.304 7.83
201.695 13.83
202.086 10.83
202.477 3.83
202.868 7.83
203.259 10.83
203.65 10.83
204.04 15.83
204.431 9.83
204.822 17.83
205.212 3.83
205.603 16.83
205.993 12.83
206.384 8.83
206.774 8.83
207.165 8.83
207.555 8.83
207.946 17.83
208.336 46.83
208.726 14.83
209.117 17.83
209.507 10.83
209.897 12.83
210.287 11.83
210.678 9.83
211.068 10.83
211.458 11.83
211.848 11.83
212.238 4.83
212.628 9.83
213.018 2.83
213.408 41.83
213.798 15.83
214.187 10.83
214.577 10.83
214.967 11.83
215.357 4.83
215.747 11.83
216.136 9.83
216.526 11.83
216.915 16.83
217.305 13.83
217.695 12.83
218.084 7.83
218.474 9.83
218.863 12.83
219.252 10.83
219.642 8.83
220.031 13.83
220.42 12.83
220.81 10.83
221.199 8.83
221.588 16.83
221.977 10.83
222.366 14.83
222.756 8.83
223.145 12.83
223.534 8.83
223.923 9.83
224.312 7.83
224.7 8.83
225.089 10.83
225.478 12.83
225.867 3.83
226.256 11.83
226.645 7.83
227.033 7.83
227.422 11.83
227.811 10.83
228.199 6.83
228.588 20.83
228.976 12.83
229.365 14.83
229.753 7.83
230.142 22.83
230.53 13.83
230.919 5.83
231.307 6.83
231.695 14.83
232.084 13.83
232.472 15.83
232.86 21.83
233.248 13.83
233.636 10.83
234.024 8.83
234.412 8.83
234.8 13.83
235.188 27.83
235.576 12.83
235.964 9.83
236.352 9.83
236.74 11.83
237.128 2.83
237.516 13.83
237.903 10.83
238.291 10.83
238.679 7.83
239.066 10.83
239.454 12.83
239.842 13.83
240.229 17.83
240.617 12.83
241.004 12.83
241.392 13.83
241.779 35.83
242.166 25.83
242.554 13.83
242.941 12.83
243.328 14.83
243.715 12.83
244.103 9.83
244.49 32.83
244.877 8.83
245.264 9.83
245.651 10.83
246.038 14.83
246.425 10.83
246.812 10.83
247.199 10.83
247.586 14.83
247.973 17.83
248.359 10.83
248.746 10.83
249.133 10.83
249.52 13.83
249.906 9.83
250.293 14.83
250.68 13.83
251.066 14.83
251.453 8.83
251.839 14.83
252.226 9.83
252.612 15.83
252.998 11.83
253.385 11.83
253.771 10.83
254.157 89.83
254.544 11.83
254.93 11.83
255.316 6.83
255.702 13.83
256.088 18.83
256.474 4.83
256.86 6.83
257.246 10.83
257.632 11.83
258.018 10.83
258.404 12.83
258.79 8.83
259.176 34.83
259.561 11.83
259.947 16.83
260.333 14.83
260.719 12.83
261.104 13.83
261.49 29.83
261.875 12.83
262.261 12.83
262.646 11.83
263.032 10.83
263.417 44.83
263.803 14.83
264.188 11.83
264.573 6.83
264.959 19.83
265.344 10.83
265.729 13.83
266.114 2.83
266.499 10.83
266.885 7.83
267.27 14.83
267.655 15.83
268.04 6.83
268.425 12.83
268.81 15.83
269.194 11.83
269.579 44.83
269.964 19.83
270.349 15.83
270.734 13.83
271.118 11.83
271.503 9.83
271.888 13.83
272.272 38.83
272.657 11.83
273.041 13.83
273.426 12.83
273.81 11.83
274.195 12.83
274.579 12.83
274.964 15.83
275.348 5.83
275.732 15.83
276.116 9.83
276.501 8.83
276.885 13.83
277.269 7.83
277.653 14.83
278.037 12.83
278.421 16.83
278.805 10.83
279.189 13.83
279.573 13.83
279.957 17.83
280.341 12.83
280.724 16.83
281.108 17.83
281.492 7.83
281.876 15.83
282.259 15.83
282.643 9.83
283.027 9.83
283.41 15.83
283.794 19.83
284.177 25.83
284.561 13.83
284.944 9.83
285.327 11.83
285.711 8.83
286.094 7.83
286.477 11.83
286.861 10.83
287.244 17.83
287.627 12.83
288.01 8.83
288.393 14.83
288.776 12.83
289.159 8.83
289.542 9.83
289.925 51.83
290.308 18.83
290.691 18.83
291.074 14.83
291.457 10.83
291.839 6.83
292.222 8.83
292.605 12.83
292.987 14.83
293.37 31.83
293.753 11.83
294.135 13.83
294.518 10.83
294.9 9.83
295.282 6.83
295.665 11.83
296.047 7.83
296.43 11.83
296.812 4.83
297.194 8.83
297.576 13.83
297.958 16.83
298.341 4.83
298.723 10.83
299.105 9.83
299.487 11.83
299.869 15.83
300.251 12.83
300.633 18.83
301.015 12.83
301.396 14.83
301.778 9.83
302.16 11.83
302.542 9.83
302.923 10.83
303.305 11.83
303.687 37.83
304.068 10.83
304.45 17.83
304.831 6.83
305.213 14.83
305.594 12.83
305.976 10.83
306.357 6.83
306.738 16.83
307.12 8.83
307.501 8.83
307.882 8.83
308.263 17.83
308.644 7.83
309.026 15.83
309.407 16.83
309.788 8.83
310.169 16.83
310.55 12.83
310.931 11.83
311.311 10.83
311.692 9.83
312.073 17.83
312.454 18.83
312.835 15.83
313.215 16.83
313.596 19.83
313.977 10.83
314.357 20.83
314.738 9.83
315.118 15.83
315.499 19.83
315.879 21.83
316.26 5.83
316.64 11.83
317.02 8.83
317.401 8.83
317.781 16.83
318.161 26.83
318.541 8.83
318.921 12.83
319.301 9.83
319.682 19.83
320.062 3.83
320.442 10.83
320.822 10.83
321.201 15.83
321.581 16.83
321.961 7.83
322.341 12.83
322.721 13.83
323.1 16.83
323.48 11.83
323.86 13.83
324.239 10.83
324.619 11.83
324.999 11.83
325.378 13.83
325.758 14.83
326.137 6.83
326.516 16.83
326.896 11.83
327.275 12.83
327.654 12.83
328.034 11.83
328.413 11.83
328.792 10.83
329.171 13.83
329.55 12.83
329.929 9.83
330.308 13.83
330.687 20.83
331.066 21.83
331.445 16.83
331.824 10.83
332.203 14.83
332.582 13.83
332.96 9.83
333.339 68.83
333.718 12.83
334.096 9.83
334.475 4.83
334.854 10.83
335.232 9.83
335.611 12.83
335.989 4.83
336.367 15.83
336.746 11.83
337.124 10.83
337.502 13.83
337.881 6.83
338.259 13.83
338.637 32.83
339.015 13.83
339.393 12.83
339.771 15.83
340.149 11.83
340.527 15.83
340.905 10.83
341.283 14.83
341.661 11.83
342.039 12.83
342.417 53.83
342.795 11.83
343.172 10.83
343.55 12.83
343.928 15.83
344.305 11.83
344.683 14.83
345.06 6.83
345.438 9.83
345.815 19.83
346.193 18.83
346.57 8.83
346.948 12.83
347.325 14.83
347.702 8.83
348.079 11.83
348.457 45.83
348.834 10.83
349.211 10.83
349.588 13.83
349.965 10.83
350.342 15.83
350.719 10.83
351.096 11.83
351.473 14.83
351.85 21.83
352.226 39.83
352.603 16.83
352.98 16.83
353.357 14.83
353.733 15.83
354.11 10.83
354.487 5.83
354.863 14.83
355.24 11.83
355.616 13.83
355.992 11.83
356.369 10.83
356.745 11.83
357.122 19.83
357.498 10.83
357.874 13.83
358.25 7.83
358.626 6.83
359.003 10.83
359.379 8.83
359.755 10.83
360.131 32.83
360.507 12.83
360.883 9.83
361.258 13.83
361.634 12.83
362.01 12.83
362.386 15.83
362.762 14.83
363.137 12.83
363.513 6.83
363.889 6.83
364.264 9.83
364.64 14.83
365.015 10.83
365.391 10.83
365.766 11.83
366.142 8.83
366.517 15.83
366.892 11.83
367.267 11.83
367.643 16.83
368.018 18.83
368.393 15.83
368.768 39.83
369.143 17.83
369.518 18.83
369.893 18.83
370.268 12.83
370.643 12.83
371.018 19.83
371.393 14.83
371.768 18.83
372.142 23.83
372.517 19.83
372.892 31.83
373.266 15.83
373.641 16.83
374.016 18.83
374.39 13.83
374.765 18.83
375.139 24.83
375.514 26.83
375.888 11.83
376.262 25.83
376.637 21.83
377.011 22.83
377.385 21.83
377.759 20.83
378.133 23.83
378.507 32.83
378.882 33.83
379.256 25.83
379.63 23.83
380.003 26.83
380.377 23.83
380.751 29.83
381.125 27.83
381.499 25.83
381.873 28.83
382.246 27.83
382.62 24.83
382.994 27.83
383.367 25.83
383.741 31.83
384.114 35.83
384.488 33.83
384.861 31.83
385.235 22.83
385.608 31.83
385.981 32.83
386.355 18.83
386.728 25.83
387.101 33.83
387.474 28.83
387.847 27.83
388.22 36.83
388.593 33.83
388.966 36.83
389.339 34.83
389.712 31.83
390.085 33.83
390.458 30.83
390.831 28.83
391.204 34.83
391.576 29.83
391.949 28.83
392.322 29.83
392.694 29.83
393.067 31.83
393.439 33.83
393.812 31.83
394.184 39.83
394.557 47.83
394.929 39.83
395.301 220.83
395.674 31.83
396.046 34.83
396.418 33.83
396.79 36.83
397.162 38.83
397.534 32.83
397.906 41.83
398.278 38.83
398.65 42.83
399.022 31.83
399.394 40.83
399.766 41.83
400.138 35.83
400.51 40.83
400.881 34.83
401.253 35.83
401.625 26.83
401.996 25.83
402.368 28.83
402.739 25.83
403.111 35.83
403.482 21.83
403.854 25.83
404.225 25.83
404.597 25.83
404.968 46.83
405.339 36.83
405.71 41.83
406.081 47.83
406.453 47.83
406.824 47.83
407.195 48.83
407.566 41.83
407.937 43.83
408.308 45.83
408.679 52.83
409.049 44.83
409.42 52.83
409.791 57.83
410.162 46.83
410.532 52.83
410.903 51.83
411.274 52.83
411.644 55.83
412.015 51.83
412.385 53.83
412.756 50.83
413.126 41.83
413.496 41.83
413.867 47.83
414.237 36.83
414.607 37.83
414.977 34.83
415.348 32.83
415.718 54.83
416.088 31.83
416.458 32.83
416.828 38.83
417.198 43.83
417.568 47.83
417.938 49.83
418.308 54.83
418.677 60.83
419.047 70.83
419.417 73.83
419.786 74.83
420.156 74.83
420.526 71.83
420.895 72.83
421.265 66.83
421.634 72.83
422.004 73.83
422.373 77.83
422.742 81.83
423.112 90.83
423.481 109.83
423.85 96.83
424.219 85.83
424.589 80.83
424.958 80.83
425.327 74.83
425.696 72.83
426.065 82.83
426.434 68.83
426.803 60.83
427.172 61.83
427.54 51.83
427.909 48.83
428.278 59.83
428.647 73.83
429.015 67.83
429.384 81.83
429.753 70.83
430.121 100.83
430.49 80.83
430.858 96.83
431.226 93.83
431.595 97.83
431.963 101.83
432.331 98.83
432.7 93.83
433.068 94.83
433.436 88.83
433.804 85.83
434.172 97.83
434.54 97.83
434.908 76.83
435.276 87.83
435.644 90.83
436.012 78.83
436.38 82.83
436.748 89.83
437.116 94.83
437.483 127.83
437.851 123.83
438.219 106.83
438.586 107.83
438.954 105.83
439.321 101.83
439.689 104.83
440.056 103.83
440.424 109.83
440.791 104.83
441.158 101.83
441.526 95.83
441.893 94.83
442.26 91.83
442.627 74.83
442.994 90.83
443.361 96.83
443.728 113.83
444.095 132.83
444.462 127.83
444.829 153.83
445.196 138.83
445.563 139.83
445.93 122.83
446.296 108.83
446.663 96.83
447.03 84.83
447.396 82.83
447.763 70.83
448.129 64.83
448.496 63.83
448.862 72.83
449.229 73.83
449.595 88.83
449.961 98.83
450.328 97.83
450.694 110.83
451.06 101.83
451.426 79.83
451.792 81.83
452.159 61.83
452.525 57.83
452.891 45.83
453.256 34.83
453.622 31.83
453.988 39.83
454.354 28.83
454.72 26.83
455.086 28.83
455.451 21.83
455.817 29.83
456.183 32.83
456.548 31.83
456.914 31.83
457.279 33.83
457.645 31.83
458.01 34.83
458.375 29.83
458.741 26.83
459.106 31.83
459.471 20.83
459.836 29.83
460.202 22.83
460.567 27.83
460.932 20.83
461.297 18.83
461.662 32.83
462.027 23.83
462.392 15.83
462.757 16.83
463.121 20.83
463.486 17.83
463.851 20.83
464.216 20.83
464.58 16.83
464.945 22.83
465.309 18.83
465.674 21.83
466.038 17.83
466.403 19.83
466.767 18.83
467.132 31.83
467.496 18.83
467.86 16.83
468.224 20.83
468.589 23.83
468.953 17.83
469.317 18.83
469.681 19.83
470.045 28.83
470.409 19.83
470.773 23.83
471.137 22.83
471.501 26.83
471.864 55.83
472.228 51.83
472.592 10.83
472.956 22.83
473.319 37.83
473.683 23.83
474.046 22.83
474.41 30.83
474.773 22.83
475.137 23.83
475.5 28.83
475.864 22.83
476.227 27.83
476.59 21.83
476.953 25.83
477.317 25.83
477.68 33.83
478.043 47.83
478.406 25.83
478.769 25.83
479.132 25.83
479.495 26.83
479.858 31.83
480.221 24.83
480.583 29.83
480.946 24.83
481.309 29.83
481.671 25.83
482.034 30.83
482.397 27.83
482.759 31.83
483.122 26.83
483.484 26.83
483.847 28.83
484.209 34.83
484.571 36.83
484.934 35.83
485.296 42.83
485.658 41.83
486.02 64.83
486.382 76.83
486.744 97.83
487.106 126.83
487.468 164.83
487.83 213.83
488.192 263.83
488.554 319.83
488.916 402.83
489.278 459.83
489.639 484.83
490.001 517.83
490.363 561.83
490.724 587.83
491.086 597.83
491.447 635.83
491.809 676.83
492.17 702.83
492.532 765.83
492.893 792.83
493.254 888.83
493.616 986.83
493.977 1111.83
494.338 1213.83
494.699 1335.83
495.06 1426.83
495.421 1546.83
495.782 1621.83
496.143 1639.83
496.504 1722.83
496.865 1760.83
497.226 1796.83
497.587 1841.83
497.948 1889.83
498.308 1877.83
498.669 1905.83
499.03 1963.83
499.39 1945.83
499.751 1935.83
500.111 1953.83
500.472 2013.83
500.832 2007.83
501.192 1998.83
501.553 2029.83
501.913 2088.83
502.273 2055.83
502.633 2078.83
502.993 2088.83
503.354 2104.83
503.714 2133.83
504.074 2160.83
504.434 2147.83
504.794 2160.83
505.153 2190.83
505.513 2201.83
505.873 2199.83
506.233 2232.83
506.593 2217.83
506.952 2243.83
507.312 2261.83
507.671 2295.83
508.031 2298.83
508.39 2310.83
508.75 2330.83
509.109 2334.83
509.469 2362.83
509.828 2341.83
510.187 2385.83
510.547 2386.83
510.906 2385.83
511.265 2445.83
511.624 2404.83
511.983 2441.83
512.342 2479.83
512.701 2499.83
513.06 2475.83
513.419 2506.83
513.778 2514.83
514.136 2544.83
514.495 2544.83
514.854 2529.83
515.213 2583.83
515.571 2559.83
515.93 2597.83
516.288 2639.83
516.647 2618.83
517.005 2696.83
517.364 2665.83
517.722 2708.83
518.08 2732.83
518.439 2744.83
518.797 2771.83
519.155 2760.83
519.513 2816.83
519.871 2807.83
520.229 2847.83
520.587 2821.83
520.945 2840.83
521.303 2909.83
521.661 2938.83
522.019 2913.83
522.377 2911.83
522.735 2940.83
523.092 2956.83
523.45 2952.83
523.808 2984.83
524.165 2993.83
524.523 3012.83
524.88 3009.83
525.238 3016.83
525.595 3036.83
525.952 3031.83
526.31 3030.83
526.667 3058.83
527.024 3084.83
527.381 3077.83
527.738 3106.83
528.096 3119.83
528.453 3085.83
528.81 3111.83
529.167 3145.83
529.524 3104.83
529.88 3128.83
530.237 3058.83
530.594 2987.83
530.951 2954.83
531.307 2846.83
531.664 2756.83
532.021 2615.83
532.377 2459.83
532.734 2314.83
533.09 2149.83
533.447 2005.83
533.803 1868.83
534.159 1681.83
534.516 1587.83
534.872 1400.83
535.228 1285.83
535.584 1180.83
535.94 1048.83
536.297 956.83
536.653 892.83
537.009 816.83
537.365 735.83
537.72 696.83
538.076 652.83
538.432 593.83
538.788 564.83
539.144 515.83
539.499 484.83
539.855 453.83
540.21 432.83
540.566 411.83
540.922 400.83
541.277 388.83
541.632 381.83
541.988 364.83
542.343 351.83
542.698 344.83
543.054 333.83
543.409 340.83
543.764 340.83
544.119 358.83
544.474 338.83
544.829 337.83
545.184 342.83
545.539 337.83
545.894 359.83
546.249 377.83
546.604 382.83
546.958 399.83
547.313 423.83
547.668 449.83
548.022 492.83
548.377 498.83
548.731 547.83
549.086 599.83
549.44 665.83
549.795 713.83
550.149 777.83
550.503 860.83
550.858 946.83
551.212 1062.83
551.566 1170.83
551.92 1276.83
552.274 1407.83
552.628 1498.83
552.982 1628.83
553.336 1725.83
553.69 1830.83
554.044 1864.83
554.398 1930.83
554.751 1948.83
555.105 1915.83
555.459 1819.83
555.812 1744.83
556.166 1720.83
556.519 1538.83
556.873 1441.83
557.226 1369.83
557.58 1232.83
557.933 1147.83
558.286 1049.83
558.64 979.83
558.993 905.83
559.346 859.83
559.699 801.83
560.052 772.83
560.405 800.83
560.758 732.83
561.111 725.83
561.464 700.83
561.817 696.83
562.17 692.83
562.523 701.83
562.875 707.83
563.228 701.83
563.581 733.83
563.933 726.83
564.286 775.83
564.638 828.83
564.991 876.83
565.343 882.83
565.696 904.83
566.048 971.83
566.4 1010.83
566.752 1097.83
567.105 1160.83
567.457 1226.83
567.809 1303.83
568.161 1404.83
568.513 1500.83
568.865 1627.83
569.217 1738.83
569.569 1872.83
569.92 1983.83
570.272 2149.83
570.624 2288.83
570.976 2413.83
571.327 2550.83
571.679 2715.83
572.03 2847.83
572.382 3017.83
572.733 3121.83
573.085 3226.83
573.436 3281.83
573.787 3379.83
574.139 3424.83
574.49 3425.83
574.841 3426.83
575.192 3462.83
575.543 3476.83
575.894 3479.83
576.245 3443.83
576.596 3439.83
576.947 3400.83
577.298 3409.83
577.649 3395.83
578 3391.83
578.35 3389.83
578.701 3348.83
579.052 3401.83
579.402 3387.83
579.753 3394.83
580.103 3393.83
580.454 3425.83
580.804 3432.83
581.155 3438.83
581.505 3487.83
581.855 3488.83
582.205 3531.83
582.556 3492.83
582.906 3495.83
583.256 3495.83
583.606 3488.83
583.956 3476.83
584.306 3493.83
584.656 3482.83
585.005 3413.83
585.355 3403.83
585.705 3381.83
586.055 3311.83
586.404 3303.83
586.754 3314.83
587.104 3300.83
587.453 3214.83
587.803 3226.83
588.152 3189.83
588.502 3177.83
588.851 3107.83
589.2 3123.83
589.549 3113.83
589.899 3114.83
590.248 3099.83
590.597 3086.83
590.946 3115.83
591.295 3089.83
591.644 3166.83
591.993 3097.83
592.342 3112.83
592.691 3148.83
593.039 3147.83
593.388 3173.83
593.737 3183.83
594.086 3202.83
594.434 3216.83
594.783 3215.83
595.131 3238.83
595.48 3287.83
595.828 3271.83
596.177 3286.83
596.525 3311.83
596.873 3336.83
597.221 3338.83
597.57 3364.83
597.918 3397.83
598.266 3375.83
598.614 3380.83
598.962 3397.83
599.31 3384.83
599.658 3381.83
600.006 3416.83
600.354 3392.83
600.701 3407.83
601.049 3397.83
601.397 3361.83
601.744 3354.83
602.092 3328.83
602.439 3335.83
602.787 3322.83
603.134 3296.83
603.482 3262.83
603.829 3224.83
604.177 3241.83
604.524 3149.83
604.871 3134.83
605.218 3063.83
605.565 3095.83
605.912 3058.83
606.259 3003.83
606.606 2967.83
606.953 2983.83
607.3 2915.83
607.647 2876.83
607.994 2874.83
608.341 2828.83
608.687 2794.83
609.034 2785.83
609.381 2751.83
609.727 2706.83
610.074 2716.83
610.42 2657.83
610.767 2664.83
611.113 2655.83
611.459 2610.83
611.806 2623.83
612.152 2591.83
612.498 2637.83
612.844 2584.83
613.19 2577.83
613.536 2589.83
613.882 2544.83
614.228 2582.83
614.574 2539.83
614.92 2573.83
615.266 2547.83
615.612 2568.83
615.957 2556.83
616.303 2589.83
616.649 2572.83
616.994 2566.83
617.34 2571.83
617.685 2589.83
618.031 2556.83
618.376 2557.83
618.721 2564.83
619.067 2531.83
619.412 2554.83
619.757 2542.83
620.102 2517.83
620.448 2483.83
620.793 2494.83
621.138 2477.83
621.483 2434.83
621.827 2445.83
622.172 2437.83
622.517 2421.83
622.862 2384.83
623.207 2377.83
623.551 2366.83
623.896 2373.83
624.241 2304.83
624.585 2314.83
624.93 2299.83
625.274 2294.83
625.619 2252.83
625.963 2275.83
626.307 2217.83
626.651 2262.83
626.996 2241.83
627.34 2255.83
627.684 2257.83
628.028 2248.83
628.372 2322.83
628.716 2279.83
629.06 2333.83
629.404 2323.83
629.748 2373.83
630.091 2386.83
630.435 2429.83
630.779 2442.83
631.123 2481.83
631.466 2530.83
631.81 2562.83
632.153 2588.83
632.497 2635.83
632.84 2657.83
633.183 2719.83
633.527 2774.83
633.87 2804.83
634.213 2832.83
634.556 2874.83
634.9 2867.83
635.243 2907.83
635.586 2860.83
635.929 2873.83
636.272 2859.83
636.614 2860.83
636.957 2860.83
637.3 2808.83
637.643 2760.83
637.986 2711.83
638.328 2671.83
638.671 2647.83
639.013 2611.83
639.356 2557.83
639.698 2505.83
640.041 2468.83
640.383 2400.83
640.725 2338.83
641.068 2315.83
641.41 2235.83
641.752 2204.83
642.094 2131.83
642.436 2106.83
642.778 2070.83
643.12 2028.83
643.462 2013.83
643.804 2000.83
644.146 1943.83
644.488 1948.83
644.829 1908.83
645.171 1882.83
645.513 1915.83
645.854 1861.83
646.196 1878.83
646.537 1899.83
646.879 1893.83
647.22 1909.83
647.562 1934.83
647.903 1925.83
648.244 1994.83
648.585 2018.83
648.926 2066.83
649.268 2064.83
649.609 2124.83
649.95 2182.83
650.291 2249.83
650.632 2286.83
650.972 2326.83
651.313 2393.83
651.654 2438.83
651.995 2514.83
652.335 2536.83
652.676 2529.83
653.017 2588.83
653.357 2611.83
653.698 2610.83
654.038 2638.83
654.378 2554.83
654.719 2623.83
655.059 2524.83
655.399 2530.83
655.74 2480.83
656.08 2415.83
656.42 2406.83
656.76 2385.83
657.1 2336.83
657.44 2259.83
657.78 2199.83
658.12 2108.83
658.459 2111.83
658.799 2031.83
659.139 1962.83
659.479 1940.83
659.818 1900.83
660.158 1867.83
660.497 1796.83
660.837 1728.83
661.176 1694.83
661.516 1654.83
661.855 1619.83
662.194 1575.83
662.533 1570.83
662.873 1512.83
663.212 1509.83
663.551 1474.83
663.89 1439.83
664.229 1425.83
664.568 1424.83
664.907 1428.83
665.245 1412.83
665.584 1404.83
665.923 1407.83
666.262 1412.83
666.6 1455.83
666.939 1434.83
667.277 1466.83
667.616 1486.83
667.954 1527.83
668.293 1539.83
668.631 1564.83
668.969 1629.83
669.308 1711.83
669.646 1748.83
669.984 1803.83
670.322 1853.83
670.66 1944.83
670.998 1988.83
671.336 2042.83
671.674 2130.83
672.012 2177.83
672.35 2233.83
672.688 2267.83
673.025 2335.83
673.363 2335.83
673.701 2340.83
674.038 2362.83
674.376 2349.83
674.713 2323.83
675.051 2313.83
675.388 2269.83
675.725 2240.83
676.063 2232.83
676.4 2186.83
676.737 2138.83
677.074 2110.83
677.411 2071.83
677.748 2020.83
678.085 1972.83
678.422 1918.83
678.759 1918.83
679.096 1867.83
679.433 1822.83
679.769 1805.83
680.106 1752.83
680.443 1730.83
680.779 1742.83
681.116 1661.83
681.452 1647.83
681.789 1623.83
682.125 1611.83
682.461 1593.83
682.798 1589.83
683.134 1579.83
683.47 1580.83
683.806 1549.83
684.142 1586.83
684.478 1591.83
684.814 1624.83
685.15 1636.83
685.486 1638.83
685.822 1691.83
686.158 1702.83
686.494 1753.83
686.829 1819.83
687.165 1878.83
687.501 1899.83
687.836 1954.83
688.172 2031.83
688.507 2058.83
688.843 2099.83
689.178 2104.83
689.513 2167.83
689.848 2111.83
690.184 2129.83
690.519 2082.83
690.854 2087.83
691.189 2032.83
691.524 1973.83
691.859 1978.83
692.194 1920.83
692.529 1900.83
692.864 1862.83
693.198 1774.83
693.533 1732.83
693.868 1707.83
694.202 1679.83
694.537 1645.83
694.871 1613.83
695.206 1576.83
695.54 1536.83
695.875 1499.83
696.209 1437.83
696.543 1424.83
696.877 1373.83
697.212 1298.83
697.546 1294.83
697.88 1232.83
698.214 1199.83
698.548 1150.83
698.882 1106.83
699.215 1051.83
699.549 1022.83
699.883 979.83
700.217 937.83
700.55 900.83
700.884 871.83
701.218 832.83
701.551 801.83
701.885 783.83
702.218 744.83
702.551 724.83
702.885 703.83
703.218 677.83
703.551 656.83
703.884 644.83
704.217 634.83
704.551 618.83
704.884 600.83
705.217 594.83
705.549 593.83
705.882 564.83
706.215 573.83
706.548 558.83
706.881 557.83
707.213 564.83
707.546 539.83
707.878 564.83
708.211 553.83
708.543 564.83
708.876 563.83
709.208 566.83
709.541 568.83
709.873 576.83
710.205 578.83
710.537 598.83
710.869 583.83
711.201 597.83
711.533 604.83
711.865 678.83
712.197 648.83
712.529 684.83
712.861 683.83
713.193 715.83
713.524 743.83
713.856 751.83
714.188 785.83
714.519 797.83
714.851 848.83
715.182 874.83
715.514 893.83
715.845 926.83
716.176 958.83
716.507 1000.83
716.839 1045.83
717.17 1044.83
717.501 1104.83
717.832 1130.83
718.163 1161.83
718.494 1203.83
718.825 1226.83
719.156 1256.83
719.487 1314.83
719.817 1306.83
720.148 1332.83
720.479 1349.83
720.809 1369.83
721.14 1400.83
721.47 1415.83
721.801 1412.83
722.131 1436.83
722.461 1447.83
722.792 1433.83
723.122 1458.83
723.452 1466.83
723.782 1482.83
724.112 1466.83
724.442 1483.83
724.772 1473.83
725.102 1428.83
725.432 1444.83
725.762 1448.83
726.092 1437.83
726.422 1441.83
726.751 1390.83
727.081 1373.83
727.411 1388.83
727.74 1344.83
728.07 1345.83
728.399 1326.83
728.728 1294.83
729.058 1291.83
729.387 1257.83
729.716 1275.83
730.045 1210.83
730.375 1208.83
730.704 1187.83
731.033 1175.83
731.362 1143.83
731.691 1126.83
732.019 1089.83
732.348 1083.83
732.677 1071.83
733.006 1039.83
733.334 1013.83
733.663 1021.83
733.992 998.83
734.32 982.83
734.649 959.83
734.977 969.83
735.305 908.83
735.634 911.83
735.962 886.83
736.29 922.83
736.618 885.83
736.946 856.83
737.274 848.83
737.602 839.83
737.93 840.83
738.258 862.83
738.586 826.83
738.914 808.83
739.242 821.83
739.569 816.83
739.897 796.83
740.225 780.83
740.552 807.83
740.88 774.83
741.207 790.83
741.534 781.83
741.862 767.83
742.189 768.83
742.516 786.83
742.843 812.83
743.171 785.83
743.498 817.83
743.825 814.83
744.152 823.83
744.479 853.83
744.805 861.83
745.132 851.83
745.459 887.83
745.786 911.83
746.112 907.83
746.439 909.83
746.766 933.83
747.092 928.83
747.419 947.83
747.745 943.83
748.071 926.83
748.398 923.83
748.724 930.83
749.05 904.83
749.376 892.83
749.702 872.83
750.028 876.83
750.354 877.83
750.68 848.83
751.006 854.83
751.332 838.83
751.658 856.83
751.984 842.83
752.309 842.83
752.635 838.83
752.96 824.83
753.286 813.83
753.611 834.83
753.937 814.83
754.262 919.83
754.588 820.83
754.913 817.83
755.238 819.83
755.563 828.83
755.888 796.83
756.213 813.83
756.538 787.83
756.863 813.83
757.188 788.83
757.513 800.83
757.838 787.83
758.163 773.83
758.487 758.83
758.812 755.83
759.137 750.83
759.461 744.83
759.786 751.83
760.11 723.83
760.435 720.83
760.759 681.83
761.083 686.83
761.407 684.83
761.732 685.83
762.056 677.83
762.38 661.83
762.704 656.83
763.028 650.83
763.352 655.83
763.676 630.83
763.999 636.83
764.323 638.83
764.647 651.83
764.97 664.83
765.294 666.83
765.618 656.83
765.941 675.83
766.265 685.83
766.588 682.83
766.911 681.83
767.235 695.83
767.558 703.83
767.881 713.83
768.204 711.83
768.527 706.83
768.85 714.83
769.173 700.83
769.496 702.83
769.819 698.83
770.142 685.83
770.465 654.83
770.787 645.83
771.11 635.83
771.433 612.83
771.755 597.83
772.078 598.83
772.4 597.83
772.723 591.83
773.045 597.83
773.367 605.83
773.689 568.83
774.012 570.83
774.334 551.83
774.656 559.83
774.978 604.83
775.3 555.83
775.622 583.83
775.944 582.83
776.266 571.83
776.587 560.83
776.909 598.83
777.231 573.83
777.552 605.83
777.874 581.83
778.195 594.83
778.517 614.83
778.838 590.83
779.16 623.83
779.481 614.83
779.802 621.83
780.123 610.83
780.444 614.83
780.766 619.83
781.087 617.83
781.408 611.83
781.728 619.83
782.049 591.83
782.37 592.83
782.691 597.83
783.012 598.83
783.332 595.83
783.653 582.83
783.973 598.83
784.294 599.83
784.614 589.83
784.935 589.83
785.255 600.83
785.575 602.83
785.896 599.83
786.216 604.83
786.536 598.83
786.856 572.83
787.176 574.83
787.496 565.83
787.816 536.83
788.136 534.83
788.456 507.83
788.775 489.83
789.095 462.83
789.415 443.83
789.734 417.83
790.054 398.83
790.373 396.83
790.693 371.83
791.012 363.83
791.332 350.83
791.651 347.83
791.97 334.83
792.289 319.83
792.608 322.83
792.928 305.83
793.247 303.83
793.566 313.83
793.884 308.83
794.203 298.83
794.522 301.83
794.841 297.83
795.16 299.83
795.478 288.83
795.797 293.83
796.115 314.83
796.434 316.83
796.752 309.83
797.071 323.83
797.389 316.83
797.707 310.83
798.025 330.83
798.344 343.83
798.662 330.83
798.98 350.83
799.298 361.83
799.616 378.83
799.934 420.83
800.251 403.83
800.569 392.83
800.887 421.83
801.205 420.83
801.522 434.83
801.84 437.83
802.157 452.83
802.475 473.83
802.792 475.83
803.11 485.83
803.427 489.83
803.744 500.83
804.061 482.83
804.378 481.83
804.696 468.83
805.013 438.83
805.33 420.83
805.647 383.83
805.963 353.83
806.28 342.83
806.597 359.83
806.914 286.83
807.23 278.83
807.547 274.83
807.864 254.83
808.18 240.83
808.497 224.83
808.813 222.83
809.129 206.83
809.446 196.83
809.762 193.83
810.078 185.83
810.394 173.83
810.71 165.83
811.026 155.83
811.342 155.83
811.658 158.83
811.974 145.83
812.29 188.83
812.605 145.83
812.921 138.83
813.237 138.83
813.552 141.83
813.868 131.83
814.183 133.83
814.499 135.83
814.814 132.83
815.129 125.83
815.445 131.83
815.76 144.83
816.075 131.83
816.39 133.83
816.705 133.83
817.02 134.83
817.335 122.83
817.65 130.83
817.965 151.83
818.28 149.83
818.594 149.83
818.909 142.83
819.224 147.83
819.538 158.83
819.853 153.83
820.167 156.83
820.481 149.83
820.796 163.83
821.11 170.83
821.424 163.83
821.739 177.83
822.053 181.83
822.367 183.83
822.681 196.83
822.995 199.83
823.309 192.83
823.622 210.83
823.936 220.83
824.25 218.83
824.564 227.83
824.877 226.83
825.191 230.83
825.504 230.83
825.818 226.83
826.131 218.83
826.445 221.83
826.758 207.83
827.071 203.83
827.384 188.83
827.698 184.83
828.011 176.83
828.324 171.83
828.637 153.83
828.95 148.83
829.262 126.83
829.575 124.83
829.888 117.83
830.201 117.83
830.513 114.83
830.826 91.83
831.138 98.83
831.451 92.83
831.763 88.83
832.076 84.83
832.388 82.83
832.7 80.83
833.013 71.83
833.325 71.83
833.637 68.83
833.949 68.83
834.261 69.83
834.573 72.83
834.885 101.83
835.197 72.83
835.508 68.83
835.82 65.83
836.132 68.83
836.443 66.83
836.755 63.83
837.066 76.83
837.378 60.83
837.689 62.83
838.001 67.83
838.312 66.83
838.623 68.83
838.934 61.83
839.245 72.83
839.556 68.83
839.867 80.83
840.178 71.83
840.489 67.83
840.8 75.83
841.111 79.83
841.422 93.83
841.732 66.83
842.043 81.83
842.353 74.83
842.664 79.83
842.974 71.83
843.285 70.83
843.595 75.83
843.905 72.83
844.216 80.83
844.526 68.83
844.836 65.83
845.146 56.83
845.456 55.83
845.766 54.83
846.076 48.83
846.386 55.83
846.696 39.83
847.005 52.83
847.315 49.83
847.625 48.83
847.934 46.83
848.244 50.83
848.553 38.83
848.863 46.83
849.172 38.83
849.481 36.83
849.79 49.83
850.1 36.83
850.409 40.83
850.718 40.83
851.027 42.83
851.336 36.83
851.645 41.83
851.954 46.83
852.262 47.83
852.571 41.83
852.88 38.83
853.188 71.83
853.497 48.83
853.806 45.83
854.114 43.83
854.422 45.83
854.731 47.83
855.039 45.83
855.347 43.83
855.655 53.83
855.964 72.83
856.272 44.83
856.58 48.83
856.888 43.83
857.196 43.83
857.503 41.83
857.811 37.83
858.119 44.83
858.427 45.83
858.734 44.83
859.042 40.83
859.349 43.83
859.657 49.83
859.964 42.83
860.272 50.83
860.579 46.83
860.886 53.83
861.193 46.83
861.5 46.83
861.807 37.83
862.115 43.83
862.421 41.83
862.728 44.83
863.035 41.83
863.342 57.83
863.649 46.83
863.955 47.83
864.262 44.83
864.569 45.83
864.875 43.83
865.182 40.83
865.488 42.83
865.794 50.83
866.101 43.83
866.407 45.83
866.713 55.83
867.019 45.83
867.325 43.83
867.631 45.83
867.937 51.83
868.243 55.83
868.549 46.83
868.855 44.83
869.16 41.83
869.466 47.83
869.772 48.83
870.077 53.83
870.383 50.83
870.688 49.83
870.994 76.83
871.299 48.83
871.604 45.83
871.909 49.83
872.215 44.83
872.52 50.83
872.825 50.83
873.13 47.83
873.435 49.83
873.74 46.83
874.044 47.83
874.349 52.83
874.654 53.83
874.959 49.83
875.263 50.83
875.568 46.83
875.872 46.83
876.177 51.83
876.481 52.83
876.785 51.83
877.09 53.83
877.394 56.83
877.698 49.83
878.002 51.83
878.306 48.83
878.61 47.83
878.914 55.83
879.218 51.83
879.522 45.83
879.825 56.83
880.129 56.83
880.433 51.83
880.736 51.83
881.04 52.83
881.343 78.83
881.647 52.83
881.95 54.83
882.253 53.83
882.557 51.83
882.86 42.83
883.163 90.83
883.466 51.83
883.769 49.83
884.072 51.83
884.375 46.83
884.678 51.83
884.981 50.83
885.283 50.83
885.586 47.83
885.889 58.83
886.191 50.83
886.494 52.83
886.796 55.83
887.099 51.83
887.401 55.83
887.703 54.83
888.005 47.83
888.308 48.83
888.61 47.83
888.912 44.83
889.214 50.83
889.516 48.83
889.818 45.83
890.119 52.83
890.421 48.83
890.723 55.83
891.024 53.83
891.326 45.83
891.628 48.83
891.929 48.83
892.231 51.83
892.532 53.83
892.833 50.83
893.134 80.83
893.436 53.83
893.737 48.83
894.038 43.83
894.339 43.83
894.64 73.83
894.941 42.83
895.242 54.83
895.542 53.83
895.843 47.83
896.144 48.83
896.445 62.83
896.745 58.83
897.046 47.83
897.346 43.83
897.646 46.83
897.947 53.83
898.247 51.83
898.547 49.83
898.847 87.83
899.148 101.83
899.448 44.83
899.748 46.83
900.048 52.83
900.347 49.83
900.647 43.83
900.947 47.83
901.247 42.83
901.546 49.83
901.846 41.83
902.146 120.83
902.445 46.83
902.744 43.83
903.044 45.83
903.343 40.83
903.642 45.83
903.942 42.83
904.241 42.83
904.54 44.83
904.839 44.83
905.138 46.83
905.437 36.83
905.736 40.83
906.034 44.83
906.333 45.83
Attachments
Last edited: