New developments:
The resistor came in last night. After work before even eating I ran out side to the garage and soldered it in as close to the tube as possible, I also used a cotton swap and some Eyeglass cleaner to clean the mirrors. Flipped the power supply on and BAM! Photons! No flicker, great output. Seems awesome. I was ecstatic this dead tube had been brought back to life!
Now the down side. I let it run for 15mins with no issues. I turned it off....a short while later I went back out to try it again and all it did was flicker like it was doing the first day
hmm.
Than I noticed my wiring from the Alden connection to the leads leading to the anode and cathode were arcing at the crimps.(less than ideal but it was ok for testing I suppose)
-I spread the +\- further apart soldered them and taped them. That mitigated the sparking but not the flicker.
- I soldered the 75k ohm ballast CLOSER to the anode <1"
- increase wire size to 18guage? Shouldn't matter too much but worth a try.
-by passing the 10k ohm on the other side.
All resulted in a flicker.
I might try changing the 75k ohm 5 Watt resistor. Could have failed.
Thought?
Is this cycling on/off and testing going to damage my lab PS? I do not want to toast that. Also what would happen if it was wiring backwards?
Thanks guys. I was so sure it was fixed I didn't even get a photo