hi,

İ was trying to measure the power of DPSS green laser using 'ilx lightwave 6810b power meter'.
The problem i am facing is that the wavelength the meter is showing is 790 nm instead of 532nm. İ am confused? Does the laser also emit the 1064nm light too?

Best Regards

Hamza
 





The strongest line is 532nm, there are smaller 808nm lines and 1064nm lines as well.
 
Indeed arg is right. The meter is probably is confused on the wavelength due to the laser not having proper filtering possibly.
 
hi,

İ was trying to measure the power of DPSS green laser using 'ilx lightwave 6810b power meter'.
The problem i am facing is that the wavelength the meter is showing is 790 nm instead of 532nm. İ am confused? Does the laser also emit the 1064nm light too?

Best Regards

Hamza

Have you any IR-cut filter? Or a spectrometer to check for 808 and 1064.
 
İ measured the spectrum with a spectrometer but there was one one peak around 532 nm. İ will take the spectrum again on monday with a better calibrated spectrometer.

Attaching file of previous spectrum

BR

Hamza
 

Attachments

İ will take the spectrum again on monday with a better calibrated spectrometer.

Just measure farther on the IR side - you stopped at 1045,29. Have at least 1070.

808 nm is from a pumping diode, and it is focused on the crystal, then it gets absorbed and whatever is left has a large divergence. Your measurements did not show anything at 808. On the other hand 1064 nm has similar properties, like a direction and beams size, as 532 nm.

I have no experience with your ILX power meter, but mine requires to enter an approximate wavelength value to get a trustworthy measurement. Can you do that with yours?
 