NOTE:
This is in no way meant to bash anything or anyone... it is an as fair as
I could make modification of the Review LPM that we paid for...
to make it slightly better with the least amount of additional costs...
We just finished a series of tests and modifications on the Kenometer Pro
that we Traded with member CDBEAM...
We did some basic tests and the results will be posted below in a
Before and After format..
When we received the Pro from CDBEAM we started by performing a few
simple tests....
- Firmware functionality
- power reading accuracy against Newport 1825C
- a simplified external noise immunity test (Cell Phone)
- Power consumption
Firmware functionality
The firmware seemed to function well. In Menu when adjusting +/-
values... pressing the Fast Button increases the incr/decr speed..
Adjusting different values for the Graph Timing and Ranges works well.
The problem is... after adjusting the display values and removing power
to the PRO... you need to re-enter your Menu settings since nothing is
stored in Memory and is lost.
Once the LCD Graph is filled the Peak and Average values displayed
no longer change values... they seem to only calculate what is on the
Graph then lock on those values until you hit the Reset Button.
Even after the Graph is filled and locked...the actual power of the Sensor
is still displayed in real time.
The Zero button MUST be pressed before taking any readings after a
Power On reboot.
There is a 1-4mW jitter on the LCD power display with no power input
to the sensor. (we solved the Jitter)
Switching from (fully charged) batteries to the USB power the Contrast
of the LCD display needed to be adjusted with every power switchover.
(we found out why)
The Zero Button Zeroes the reading even if you are at 3000mW...
Power reading accuracy
As was expected the OPHIR sensor was spot on as long as you kept
the beam of the Laser EXACTLY in the center of the 12mm active surface...
We did some tests on different areas of the 12mm active surface with a
stable Laser and you can see the variations we recorded in the pic
below...
The firmware followed the OPHIR head's output quite well... no problems
there.
Simplified external noise immunity test
We turned on a Cell Phone and called our shop to do a quick and dirty
noise test...
We physically ran the Cell phone (as close as possible) all around the
OPHIR Head... the Head Cable and all around the Pro Case with no
input to the Sensor Head. We were not impressed... See Data Below
Seeing that... we did the same tests with a Standard LaserBee I and II
(PCB only.. no Box) and the readings only dropped a max of 2mW with
no change at the Thermopile Sensor.
Power consumption
Checking the power consumption on the Power Input Jack with a 9Volt
1 Amp Switching Power supply... we noticed the PRO drawing an excessive
amount of current as far as we were concerned... The OPHIR
head only requires ~27mA...
Modifications
We decided to see if we could fix the problem with the Power Consumption...
(we fixed that... see the cmparision below)
We needed to do something about the no-input jitter without having
access to the Firmware source code... (we fixed that)
At the same time we wanted to assemble the LPM components in a
way to be able to replace any parts easily... (we fixed that)
We also wanted to reduce the Noise vulnerability as much as possible
without needing to design a special PCB.....
We noticed that the LCD.... LCD Backlight and DC/DC Converter all sucked
power from that poor little Arduino.... We also noticed a series of separarate
grounds tied anywhere.... (We fixed that)
Here are the Stats and the results of our efforts...
Original Kenometer PRO
- Power consumption ---------- 332mA
- No-Input Jitter ----------------- 1-4mW (max)
- Noise test Head --------------- +26mW (max)
- Noise test PRO ---------------- +224mW (max)
Modified to better Specs 5W LPM <---Corrected...
- Power consumption ---------- 178mA
- No-Input Jitter ----------------- 0 mW (max)
- Noise test Head --------------- +3mW (max)
- Noise test PRO ---------------- +16mW (max)
We still need to test the PC Data Logging Software... I'll update this Post when we do...
Jerry
This is in no way meant to bash anything or anyone... it is an as fair as
I could make modification of the Review LPM that we paid for...
to make it slightly better with the least amount of additional costs...
We just finished a series of tests and modifications on the Kenometer Pro
that we Traded with member CDBEAM...
We did some basic tests and the results will be posted below in a
Before and After format..
When we received the Pro from CDBEAM we started by performing a few
simple tests....
- Firmware functionality
- power reading accuracy against Newport 1825C
- a simplified external noise immunity test (Cell Phone)
- Power consumption
Firmware functionality
The firmware seemed to function well. In Menu when adjusting +/-
values... pressing the Fast Button increases the incr/decr speed..
Adjusting different values for the Graph Timing and Ranges works well.
The problem is... after adjusting the display values and removing power
to the PRO... you need to re-enter your Menu settings since nothing is
stored in Memory and is lost.
Once the LCD Graph is filled the Peak and Average values displayed
no longer change values... they seem to only calculate what is on the
Graph then lock on those values until you hit the Reset Button.
Even after the Graph is filled and locked...the actual power of the Sensor
is still displayed in real time.
The Zero button MUST be pressed before taking any readings after a
Power On reboot.
There is a 1-4mW jitter on the LCD power display with no power input
to the sensor. (we solved the Jitter)
Switching from (fully charged) batteries to the USB power the Contrast
of the LCD display needed to be adjusted with every power switchover.
(we found out why)
The Zero Button Zeroes the reading even if you are at 3000mW...
Power reading accuracy
As was expected the OPHIR sensor was spot on as long as you kept
the beam of the Laser EXACTLY in the center of the 12mm active surface...
We did some tests on different areas of the 12mm active surface with a
stable Laser and you can see the variations we recorded in the pic
below...
The firmware followed the OPHIR head's output quite well... no problems
there.
Simplified external noise immunity test
We turned on a Cell Phone and called our shop to do a quick and dirty
noise test...
We physically ran the Cell phone (as close as possible) all around the
OPHIR Head... the Head Cable and all around the Pro Case with no
input to the Sensor Head. We were not impressed... See Data Below
Seeing that... we did the same tests with a Standard LaserBee I and II
(PCB only.. no Box) and the readings only dropped a max of 2mW with
no change at the Thermopile Sensor.
Power consumption
Checking the power consumption on the Power Input Jack with a 9Volt
1 Amp Switching Power supply... we noticed the PRO drawing an excessive
amount of current as far as we were concerned... The OPHIR
head only requires ~27mA...
Modifications
We decided to see if we could fix the problem with the Power Consumption...
(we fixed that... see the cmparision below)
We needed to do something about the no-input jitter without having
access to the Firmware source code... (we fixed that)
At the same time we wanted to assemble the LPM components in a
way to be able to replace any parts easily... (we fixed that)
We also wanted to reduce the Noise vulnerability as much as possible
without needing to design a special PCB.....
We noticed that the LCD.... LCD Backlight and DC/DC Converter all sucked
power from that poor little Arduino.... We also noticed a series of separarate
grounds tied anywhere.... (We fixed that)
Here are the Stats and the results of our efforts...
Original Kenometer PRO
- Power consumption ---------- 332mA
- No-Input Jitter ----------------- 1-4mW (max)
- Noise test Head --------------- +26mW (max)
- Noise test PRO ---------------- +224mW (max)
Modified to better Specs 5W LPM <---Corrected...
- Power consumption ---------- 178mA
- No-Input Jitter ----------------- 0 mW (max)
- Noise test Head --------------- +3mW (max)
- Noise test PRO ---------------- +16mW (max)
We still need to test the PC Data Logging Software... I'll update this Post when we do...
Jerry